I would like to get any feedback or experience info on whether Deoxit
would help sufficiently to improve mechanical switch contact performance
used in very low (nV) level circuits. I posted this question in the
tekscopes group a while back, but with no response. I was surprised that
there was no info from such a wide-ranging group, but maybe here in
volt-nuts, with narrow focus, it will stand a better chance. Here is
approximately what I wrote, slightly edited, for consideration.

Then:

"One of my projects involves boxing up some special transformers and a
switching system, for isolating and dividing down AC signals by large
factors (E-3 to E-6), down as far as the 1 nV RMS region. A fair number
of switching elements are needed to route and select various transformer
output taps to send to the single output connection. There are two
transformers to cover two frequency bands 10 Hz-1 kHz, and 1 kHz-100
kHz, and each has four outputs for the decade dividing. The switching
elements need to have as low an on resistance (<< 1 ohm) as possible,
and function well at extremely low levels. Since this is an AC system,
I'm not worried about Seebeck effects, just the low level contacting
ability. The primary sides of the transformers are at more normal
levels, so the switches there are not critical.

I have all sorts of nice low level regular and Hg reed relays that would
do the job, but since the output side is to be isolated, to minimize
interference, it would be much better to do all the switching passively
with mechanical switches. This would avoid needing electricity, and the
proximity problems of having relay coils and capacitance in the low
level environment, and power supply and line noise and ground loops (no
power cord), or messing with battery power. Especially, there would be
no power transformer emissions to worry about.

I have lots of mechanical switches of all sorts. I'd like to go with a
rotary wafer type to select the transformer taps and frequency ranges. I
have mostly standard Ag plated type contact ones and some Au ones. I can
usually build a switch from pieces for almost any arrangement.

Another, but quite complicated option is to use low level reeds actuated
by a mechanically driven magnet, but doing that would be a big project
in itself, so scratch that, unless a simple way pops up. Another way is
to make heavy analog switches with big MOSFETs, but that's a lot of
parts, a battery, and issues with capacitance and crosstalk. So, good
old mechanical switching seems the best way to go - if only it can
actually be done.

The big question then, is how to get a regular, environmentally exposed
contact to be usable in the nV region. Wetting the contacts with Deoxit
is the only practical thing I can picture so far. There is another class
of contact "protectors" based on synthetic oils, but I think I need
chemical action too, so these may not work - I'll have to investigate
that too.

So anyway, does anyone have experience or knowledge of how Deoxit would
behave for very low level contacts? It may boil down to a "try it and
see" scenario, but it would be nice to have some info in advance."

Now:

I have opted for the "try it and see" scenario, and have built it up so
far using mechanical rotary wafer switches. The low-level section will
use all gold plated contacts, with Deoxit G100 applied. Also to be
included, is a "contact cleaning" mode, where some higher DC (couple
volts and mA) can be temporarily applied to all contacts from a built-in
Li cell circuit. Since I don't have enough of the right kinds of Au
plated contact wafer parts available, the higher level input (primary)
side will include some regular Ag plated contact switching too. These
will be chemically cleaned first, then have Deoxit D100 applied.

I have it mostly mechanically built, almost ready to begin wiring up and
experimenting. This will be ongoing for a while, and I would appreciate
if anyone here does have experience with such low level applications
with exposed mechanical contacts. I know we usually talk about DC here,
while this is an AC application, but the same principles apply.

Ed

I would like to get any feedback or experience info on whether Deoxit would help sufficiently to improve mechanical switch contact performance used in very low (nV) level circuits. I posted this question in the tekscopes group a while back, but with no response. I was surprised that there was no info from such a wide-ranging group, but maybe here in volt-nuts, with narrow focus, it will stand a better chance. Here is approximately what I wrote, slightly edited, for consideration. Then: "One of my projects involves boxing up some special transformers and a switching system, for isolating and dividing down AC signals by large factors (E-3 to E-6), down as far as the 1 nV RMS region. A fair number of switching elements are needed to route and select various transformer output taps to send to the single output connection. There are two transformers to cover two frequency bands 10 Hz-1 kHz, and 1 kHz-100 kHz, and each has four outputs for the decade dividing. The switching elements need to have as low an on resistance (<< 1 ohm) as possible, and function well at extremely low levels. Since this is an AC system, I'm not worried about Seebeck effects, just the low level contacting ability. The primary sides of the transformers are at more normal levels, so the switches there are not critical. I have all sorts of nice low level regular and Hg reed relays that would do the job, but since the output side is to be isolated, to minimize interference, it would be much better to do all the switching passively with mechanical switches. This would avoid needing electricity, and the proximity problems of having relay coils and capacitance in the low level environment, and power supply and line noise and ground loops (no power cord), or messing with battery power. Especially, there would be no power transformer emissions to worry about. I have lots of mechanical switches of all sorts. I'd like to go with a rotary wafer type to select the transformer taps and frequency ranges. I have mostly standard Ag plated type contact ones and some Au ones. I can usually build a switch from pieces for almost any arrangement. Another, but quite complicated option is to use low level reeds actuated by a mechanically driven magnet, but doing that would be a big project in itself, so scratch that, unless a simple way pops up. Another way is to make heavy analog switches with big MOSFETs, but that's a lot of parts, a battery, and issues with capacitance and crosstalk. So, good old mechanical switching seems the best way to go - if only it can actually be done. The big question then, is how to get a regular, environmentally exposed contact to be usable in the nV region. Wetting the contacts with Deoxit is the only practical thing I can picture so far. There is another class of contact "protectors" based on synthetic oils, but I think I need chemical action too, so these may not work - I'll have to investigate that too. So anyway, does anyone have experience or knowledge of how Deoxit would behave for very low level contacts? It may boil down to a "try it and see" scenario, but it would be nice to have some info in advance." Now: I have opted for the "try it and see" scenario, and have built it up so far using mechanical rotary wafer switches. The low-level section will use all gold plated contacts, with Deoxit G100 applied. Also to be included, is a "contact cleaning" mode, where some higher DC (couple volts and mA) can be temporarily applied to all contacts from a built-in Li cell circuit. Since I don't have enough of the right kinds of Au plated contact wafer parts available, the higher level input (primary) side will include some regular Ag plated contact switching too. These will be chemically cleaned first, then have Deoxit D100 applied. I have it mostly mechanically built, almost ready to begin wiring up and experimenting. This will be ongoing for a while, and I would appreciate if anyone here does have experience with such low level applications with exposed mechanical contacts. I know we usually talk about DC here, while this is an AC application, but the same principles apply. Ed
RB
Reginald Beardsley
Sat, Jun 6, 2020 8:28 PM

Have you considered using HP 44421A multiplexer boards from a 3497A? They are cheap, though terminal blocks can be hard to come by. They are high spec reed relays and if you get the thermocouple compensated terminal blocks you can adjust for thermal EMF.

Reg

 On Saturday, June 6, 2020, 02:15:47 PM CDT, ed breya <eb@telight.com> wrote:  

I would like to get any feedback or experience info on whether Deoxit
would help sufficiently to improve mechanical switch contact performance
used in very low (nV) level circuits. I posted this question in the
tekscopes group a while back, but with no response. I was surprised that
there was no info from such a wide-ranging group, but maybe here in
volt-nuts, with narrow focus, it will stand a better chance. Here is
approximately what I wrote, slightly edited, for consideration.

Then:

"One of my projects involves boxing up some special transformers and a
switching system, for isolating and dividing down AC signals by large
factors (E-3 to E-6), down as far as the 1 nV RMS region. A fair number
of switching elements are needed to route and select various transformer
output taps to send to the single output connection. There are two
transformers to cover two frequency bands 10 Hz-1 kHz, and 1 kHz-100
kHz, and each has four outputs for the decade dividing. The switching
elements need to have as low an on resistance (<< 1 ohm) as possible,
and function well at extremely low levels. Since this is an AC system,
I'm not worried about Seebeck effects, just the low level contacting
ability. The primary sides of the transformers are at more normal
levels, so the switches there are not critical.

I have all sorts of nice low level regular and Hg reed relays that would
do the job, but since the output side is to be isolated, to minimize
interference, it would be much better to do all the switching passively
with mechanical switches. This would avoid needing electricity, and the
proximity problems of having relay coils and capacitance in the low
level environment, and power supply and line noise and ground loops (no
power cord), or messing with battery power. Especially, there would be
no power transformer emissions to worry about.

I have lots of mechanical switches of all sorts. I'd like to go with a
rotary wafer type to select the transformer taps and frequency ranges. I
have mostly standard Ag plated type contact ones and some Au ones. I can
usually build a switch from pieces for almost any arrangement.

Another, but quite complicated option is to use low level reeds actuated
by a mechanically driven magnet, but doing that would be a big project
in itself, so scratch that, unless a simple way pops up. Another way is
to make heavy analog switches with big MOSFETs, but that's a lot of
parts, a battery, and issues with capacitance and crosstalk. So, good
old mechanical switching seems the best way to go - if only it can
actually be done.

The big question then, is how to get a regular, environmentally exposed
contact to be usable in the nV region. Wetting the contacts with Deoxit
is the only practical thing I can picture so far. There is another class
of contact "protectors" based on synthetic oils, but I think I need
chemical action too, so these may not work - I'll have to investigate
that too.

So anyway, does anyone have experience or knowledge of how Deoxit would
behave for very low level contacts? It may boil down to a "try it and
see" scenario, but it would be nice to have some info in advance."

Now:

I have opted for the "try it and see" scenario, and have built it up so
far using mechanical rotary wafer switches. The low-level section will
use all gold plated contacts, with Deoxit G100 applied. Also to be
included, is a "contact cleaning" mode, where some higher DC (couple
volts and mA) can be temporarily applied to all contacts from a built-in
Li cell circuit. Since I don't have enough of the right kinds of Au
plated contact wafer parts available, the higher level input (primary)
side will include some regular Ag plated contact switching too. These
will be chemically cleaned first, then have Deoxit D100 applied.

I have it mostly mechanically built, almost ready to begin wiring up and
experimenting. This will be ongoing for a while, and I would appreciate
if anyone here does have experience with such low level applications
with exposed mechanical contacts. I know we usually talk about DC here,
while this is an AC application, but the same principles apply.

Ed


volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.

Have you considered using HP 44421A multiplexer boards from a 3497A? They are cheap, though terminal blocks can be hard to come by. They are high spec reed relays and if you get the thermocouple compensated terminal blocks you can adjust for thermal EMF. Reg On Saturday, June 6, 2020, 02:15:47 PM CDT, ed breya <eb@telight.com> wrote: I would like to get any feedback or experience info on whether Deoxit would help sufficiently to improve mechanical switch contact performance used in very low (nV) level circuits. I posted this question in the tekscopes group a while back, but with no response. I was surprised that there was no info from such a wide-ranging group, but maybe here in volt-nuts, with narrow focus, it will stand a better chance. Here is approximately what I wrote, slightly edited, for consideration. Then: "One of my projects involves boxing up some special transformers and a switching system, for isolating and dividing down AC signals by large factors (E-3 to E-6), down as far as the 1 nV RMS region. A fair number of switching elements are needed to route and select various transformer output taps to send to the single output connection. There are two transformers to cover two frequency bands 10 Hz-1 kHz, and 1 kHz-100 kHz, and each has four outputs for the decade dividing. The switching elements need to have as low an on resistance (<< 1 ohm) as possible, and function well at extremely low levels. Since this is an AC system, I'm not worried about Seebeck effects, just the low level contacting ability. The primary sides of the transformers are at more normal levels, so the switches there are not critical. I have all sorts of nice low level regular and Hg reed relays that would do the job, but since the output side is to be isolated, to minimize interference, it would be much better to do all the switching passively with mechanical switches. This would avoid needing electricity, and the proximity problems of having relay coils and capacitance in the low level environment, and power supply and line noise and ground loops (no power cord), or messing with battery power. Especially, there would be no power transformer emissions to worry about. I have lots of mechanical switches of all sorts. I'd like to go with a rotary wafer type to select the transformer taps and frequency ranges. I have mostly standard Ag plated type contact ones and some Au ones. I can usually build a switch from pieces for almost any arrangement. Another, but quite complicated option is to use low level reeds actuated by a mechanically driven magnet, but doing that would be a big project in itself, so scratch that, unless a simple way pops up. Another way is to make heavy analog switches with big MOSFETs, but that's a lot of parts, a battery, and issues with capacitance and crosstalk. So, good old mechanical switching seems the best way to go - if only it can actually be done. The big question then, is how to get a regular, environmentally exposed contact to be usable in the nV region. Wetting the contacts with Deoxit is the only practical thing I can picture so far. There is another class of contact "protectors" based on synthetic oils, but I think I need chemical action too, so these may not work - I'll have to investigate that too. So anyway, does anyone have experience or knowledge of how Deoxit would behave for very low level contacts? It may boil down to a "try it and see" scenario, but it would be nice to have some info in advance." Now: I have opted for the "try it and see" scenario, and have built it up so far using mechanical rotary wafer switches. The low-level section will use all gold plated contacts, with Deoxit G100 applied. Also to be included, is a "contact cleaning" mode, where some higher DC (couple volts and mA) can be temporarily applied to all contacts from a built-in Li cell circuit. Since I don't have enough of the right kinds of Au plated contact wafer parts available, the higher level input (primary) side will include some regular Ag plated contact switching too. These will be chemically cleaned first, then have Deoxit D100 applied. I have it mostly mechanically built, almost ready to begin wiring up and experimenting. This will be ongoing for a while, and I would appreciate if anyone here does have experience with such low level applications with exposed mechanical contacts. I know we usually talk about DC here, while this is an AC application, but the same principles apply. Ed _______________________________________________ volt-nuts mailing list -- volt-nuts@lists.febo.com To unsubscribe, go to http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com and follow the instructions there.
DD
Dr. David Kirkby
Sat, Jun 6, 2020 9:47 PM

Dr David Kirkby Ph.D C.Eng MIET
Email: drkirkby@kirkbymicrowave.co.uk Web:
https://www.kirkbymicrowave.co.uk/
Kirkby Microwave Ltd (Tel 01621-680100 / +44 1621-680100)
Stokes Hall Lodge, Burnham Rd, Chelmsford, Essex, CM3 6DT.

On Sat, 6 Jun 2020 at 20:15, ed breya eb@telight.com wrote:

I would like to get any feedback or experience info on whether Deoxit
would help sufficiently to improve mechanical switch contact performance
used in very low (nV) level circuits. I posted this question in the
tekscopes group a while back, but with no response. I was surprised that
there was no info from such a wide-ranging group, but maybe here in
volt-nuts, with narrow focus, it will stand a better chance. Here is
approximately what I wrote, slightly edited, for consideration.

This is outside my area of expertise. I would certainly take a look at the
Keithley low-level measurement handbook

https://download.tek.com/document/LowLevelHandbook_7Ed.pdf

that might give you some ideas.

Dave

Dr David Kirkby Ph.D C.Eng MIET Email: drkirkby@kirkbymicrowave.co.uk Web: https://www.kirkbymicrowave.co.uk/ Kirkby Microwave Ltd (Tel 01621-680100 / +44 1621-680100) Stokes Hall Lodge, Burnham Rd, Chelmsford, Essex, CM3 6DT. On Sat, 6 Jun 2020 at 20:15, ed breya <eb@telight.com> wrote: > I would like to get any feedback or experience info on whether Deoxit > would help sufficiently to improve mechanical switch contact performance > used in very low (nV) level circuits. I posted this question in the > tekscopes group a while back, but with no response. I was surprised that > there was no info from such a wide-ranging group, but maybe here in > volt-nuts, with narrow focus, it will stand a better chance. Here is > approximately what I wrote, slightly edited, for consideration. > This is outside my area of expertise. I would certainly take a look at the Keithley low-level measurement handbook https://download.tek.com/document/LowLevelHandbook_7Ed.pdf that might give you some ideas. Dave
TM
Todd Micallef
Sun, Jun 7, 2020 12:14 AM

Ed,
Look up the Keithley 262 low thermal divider. The switches are made from
what appears to be unplated FR4. An article on how it was made can be found
here https://xdevs.com/fix/kei262/

The user manual describes a lubricant that should be applied to the surface
after cleaning. It happens to be made by Caig, or at least it used to be.
Perhaps they can tell you if one of their current products is the same
thing under a different name.

Todd https://xdevs.com/fix/kei262/

On Sat, Jun 6, 2020 at 5:48 PM Dr. David Kirkby <
drkirkby@kirkbymicrowave.co.uk> wrote:

Dr David Kirkby Ph.D C.Eng MIET
Email: drkirkby@kirkbymicrowave.co.uk Web:
https://www.kirkbymicrowave.co.uk/
Kirkby Microwave Ltd (Tel 01621-680100 / +44 1621-680100)
Stokes Hall Lodge, Burnham Rd, Chelmsford, Essex, CM3 6DT.

On Sat, 6 Jun 2020 at 20:15, ed breya eb@telight.com wrote:

I would like to get any feedback or experience info on whether Deoxit
would help sufficiently to improve mechanical switch contact performance
used in very low (nV) level circuits. I posted this question in the
tekscopes group a while back, but with no response. I was surprised that
there was no info from such a wide-ranging group, but maybe here in
volt-nuts, with narrow focus, it will stand a better chance. Here is
approximately what I wrote, slightly edited, for consideration.

This is outside my area of expertise. I would certainly take a look at the
Keithley low-level measurement handbook

https://download.tek.com/document/LowLevelHandbook_7Ed.pdf

that might give you some ideas.

Dave


volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.

Ed, Look up the Keithley 262 low thermal divider. The switches are made from what appears to be unplated FR4. An article on how it was made can be found here https://xdevs.com/fix/kei262/ The user manual describes a lubricant that should be applied to the surface after cleaning. It happens to be made by Caig, or at least it used to be. Perhaps they can tell you if one of their current products is the same thing under a different name. Todd <https://xdevs.com/fix/kei262/> On Sat, Jun 6, 2020 at 5:48 PM Dr. David Kirkby < drkirkby@kirkbymicrowave.co.uk> wrote: > Dr David Kirkby Ph.D C.Eng MIET > Email: drkirkby@kirkbymicrowave.co.uk Web: > https://www.kirkbymicrowave.co.uk/ > Kirkby Microwave Ltd (Tel 01621-680100 / +44 1621-680100) > Stokes Hall Lodge, Burnham Rd, Chelmsford, Essex, CM3 6DT. > > > > > > On Sat, 6 Jun 2020 at 20:15, ed breya <eb@telight.com> wrote: > > > I would like to get any feedback or experience info on whether Deoxit > > would help sufficiently to improve mechanical switch contact performance > > used in very low (nV) level circuits. I posted this question in the > > tekscopes group a while back, but with no response. I was surprised that > > there was no info from such a wide-ranging group, but maybe here in > > volt-nuts, with narrow focus, it will stand a better chance. Here is > > approximately what I wrote, slightly edited, for consideration. > > > > This is outside my area of expertise. I would certainly take a look at the > Keithley low-level measurement handbook > > https://download.tek.com/document/LowLevelHandbook_7Ed.pdf > > that might give you some ideas. > > Dave > _______________________________________________ > volt-nuts mailing list -- volt-nuts@lists.febo.com > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com > and follow the instructions there. >
EB
ed breya
Sun, Jun 7, 2020 2:07 AM

Thanks guys, for the responses already - I should have asked here in the
first place. Here's a consolidated answer:

Reg, I could use a low level scanner if it was for a general or one-time
measurement, but this is to be a compact, self-contained AC divider, for
isolated low level signal generation for experimenting with lock-in
analyzer setups. The transformer modules to be used came from a Fluke
5200A attenuator board. I initially planned on only the low and middle
band transformers, for 10 Hz to 100 kHz with good flatness, which covers
the range of all my LIAs. But then I figured I'll include the high band
(up to 1 MHz) too, using the last, little transformer. Since it's AC,
I'm not worried about thermal issues.

David, thanks for the link. I have seen and have the Keithley low level
measurements bible somewhere, but I think this is the newest edition
I've seen. I have copied the pdf into several places now.

Todd, great find! It looks like the 262 used good old copper-copper
rubbing (to cut the oxides) contacts for low level and low thermal
characteristics. Also note the lowest, final dividing resistor network
is all-copper resistance and connections. I've seen this before in old
Keithley gear. My model 148 nV meters use similar in the feedback
divider, and also heavy lug Cu-Cu switches in the front end ranging. I
see the reference to the Cramolin oils in the 262 manual - these were
indeed the forerunners of what became Deoxit. It's a long, complicated,
and kind of interesting story - I studied it a few years ago.

So anyway, I'm feeling better about going with my mechanical switches
with Au and Ag contacts and Deoxit combo. If worse comes to worse, I
guess I could scrape off the plating and go Cu all the way, but I don't
think that will be needed.

Ed

Thanks guys, for the responses already - I should have asked here in the first place. Here's a consolidated answer: Reg, I could use a low level scanner if it was for a general or one-time measurement, but this is to be a compact, self-contained AC divider, for isolated low level signal generation for experimenting with lock-in analyzer setups. The transformer modules to be used came from a Fluke 5200A attenuator board. I initially planned on only the low and middle band transformers, for 10 Hz to 100 kHz with good flatness, which covers the range of all my LIAs. But then I figured I'll include the high band (up to 1 MHz) too, using the last, little transformer. Since it's AC, I'm not worried about thermal issues. David, thanks for the link. I have seen and have the Keithley low level measurements bible somewhere, but I think this is the newest edition I've seen. I have copied the pdf into several places now. Todd, great find! It looks like the 262 used good old copper-copper rubbing (to cut the oxides) contacts for low level and low thermal characteristics. Also note the lowest, final dividing resistor network is all-copper resistance and connections. I've seen this before in old Keithley gear. My model 148 nV meters use similar in the feedback divider, and also heavy lug Cu-Cu switches in the front end ranging. I see the reference to the Cramolin oils in the 262 manual - these were indeed the forerunners of what became Deoxit. It's a long, complicated, and kind of interesting story - I studied it a few years ago. So anyway, I'm feeling better about going with my mechanical switches with Au and Ag contacts and Deoxit combo. If worse comes to worse, I guess I could scrape off the plating and go Cu all the way, but I don't think that will be needed. Ed
EB
ed breya
Sun, Jun 7, 2020 2:32 AM

One thing I should mention too is that because it's a wide bandwidth AC
system, with a wide signal range, there are issues with stray
capacitance in many places, so the internal connections and routing have
to be carefully made, and as compact as possible.

Ed

One thing I should mention too is that because it's a wide bandwidth AC system, with a wide signal range, there are issues with stray capacitance in many places, so the internal connections and routing have to be carefully made, and as compact as possible. Ed
SW
Steve Wiseman
Sun, Jun 7, 2020 9:40 AM

If you want the pre-built goodness of reed switches, without the

electrical  interference from the coils and drivers, then how about
magnets-onna-stick as your switch control element? Either push-pull like an
organ stop, or rotate on a shaft? Easy to route to a front panel, cheap,
stable, mostly good things?

Steve

> If you want the pre-built goodness of reed switches, without the electrical interference from the coils and drivers, then how about magnets-onna-stick as your switch control element? Either push-pull like an organ stop, or rotate on a shaft? Easy to route to a front panel, cheap, stable, mostly good things? Steve
DJ
Didier Juges
Sun, Jun 7, 2020 1:54 PM

It seems to me one of the problems you will fight with mechanical switches
is oxidation, or corrosion in a more general sense. It is hard
(impossible?) to make completely sealed mechanical switches.
I am not even sure that gold contacts would not eventually cause problems
at such a low level.
This is the basic reason for the development of vacuum sealed reed relays
and particularly the mercury wetted type.

Now, you only talk about establishing the contact. How about isolation? I
am not sure Deoxit is a good enough isolator in your application. It could
create leakage currents, particularly after it itself becomes contaminated.

Didier KOBB

On Sat, Jun 6, 2020 at 2:15 PM ed breya eb@telight.com wrote:

I would like to get any feedback or experience info on whether Deoxit
would help sufficiently to improve mechanical switch contact performance
used in very low (nV) level circuits. I posted this question in the
tekscopes group a while back, but with no response. I was surprised that
there was no info from such a wide-ranging group, but maybe here in
volt-nuts, with narrow focus, it will stand a better chance. Here is
approximately what I wrote, slightly edited, for consideration.

Then:

"One of my projects involves boxing up some special transformers and a
switching system, for isolating and dividing down AC signals by large
factors (E-3 to E-6), down as far as the 1 nV RMS region. A fair number
of switching elements are needed to route and select various transformer
output taps to send to the single output connection. There are two
transformers to cover two frequency bands 10 Hz-1 kHz, and 1 kHz-100
kHz, and each has four outputs for the decade dividing. The switching
elements need to have as low an on resistance (<< 1 ohm) as possible,
and function well at extremely low levels. Since this is an AC system,
I'm not worried about Seebeck effects, just the low level contacting
ability. The primary sides of the transformers are at more normal
levels, so the switches there are not critical.

I have all sorts of nice low level regular and Hg reed relays that would
do the job, but since the output side is to be isolated, to minimize
interference, it would be much better to do all the switching passively
with mechanical switches. This would avoid needing electricity, and the
proximity problems of having relay coils and capacitance in the low
level environment, and power supply and line noise and ground loops (no
power cord), or messing with battery power. Especially, there would be
no power transformer emissions to worry about.

I have lots of mechanical switches of all sorts. I'd like to go with a
rotary wafer type to select the transformer taps and frequency ranges. I
have mostly standard Ag plated type contact ones and some Au ones. I can
usually build a switch from pieces for almost any arrangement.

Another, but quite complicated option is to use low level reeds actuated
by a mechanically driven magnet, but doing that would be a big project
in itself, so scratch that, unless a simple way pops up. Another way is
to make heavy analog switches with big MOSFETs, but that's a lot of
parts, a battery, and issues with capacitance and crosstalk. So, good
old mechanical switching seems the best way to go - if only it can
actually be done.

The big question then, is how to get a regular, environmentally exposed
contact to be usable in the nV region. Wetting the contacts with Deoxit
is the only practical thing I can picture so far. There is another class
of contact "protectors" based on synthetic oils, but I think I need
chemical action too, so these may not work - I'll have to investigate
that too.

So anyway, does anyone have experience or knowledge of how Deoxit would
behave for very low level contacts? It may boil down to a "try it and
see" scenario, but it would be nice to have some info in advance."

Now:

I have opted for the "try it and see" scenario, and have built it up so
far using mechanical rotary wafer switches. The low-level section will
use all gold plated contacts, with Deoxit G100 applied. Also to be
included, is a "contact cleaning" mode, where some higher DC (couple
volts and mA) can be temporarily applied to all contacts from a built-in
Li cell circuit. Since I don't have enough of the right kinds of Au
plated contact wafer parts available, the higher level input (primary)
side will include some regular Ag plated contact switching too. These
will be chemically cleaned first, then have Deoxit D100 applied.

I have it mostly mechanically built, almost ready to begin wiring up and
experimenting. This will be ongoing for a while, and I would appreciate
if anyone here does have experience with such low level applications
with exposed mechanical contacts. I know we usually talk about DC here,
while this is an AC application, but the same principles apply.

Ed


volt-nuts mailing list -- volt-nuts@lists.febo.com
To unsubscribe, go to
http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com
and follow the instructions there.

It seems to me one of the problems you will fight with mechanical switches is oxidation, or corrosion in a more general sense. It is hard (impossible?) to make completely sealed mechanical switches. I am not even sure that gold contacts would not eventually cause problems at such a low level. This is the basic reason for the development of vacuum sealed reed relays and particularly the mercury wetted type. Now, you only talk about establishing the contact. How about isolation? I am not sure Deoxit is a good enough isolator in your application. It could create leakage currents, particularly after it itself becomes contaminated. Didier KOBB On Sat, Jun 6, 2020 at 2:15 PM ed breya <eb@telight.com> wrote: > I would like to get any feedback or experience info on whether Deoxit > would help sufficiently to improve mechanical switch contact performance > used in very low (nV) level circuits. I posted this question in the > tekscopes group a while back, but with no response. I was surprised that > there was no info from such a wide-ranging group, but maybe here in > volt-nuts, with narrow focus, it will stand a better chance. Here is > approximately what I wrote, slightly edited, for consideration. > > Then: > > "One of my projects involves boxing up some special transformers and a > switching system, for isolating and dividing down AC signals by large > factors (E-3 to E-6), down as far as the 1 nV RMS region. A fair number > of switching elements are needed to route and select various transformer > output taps to send to the single output connection. There are two > transformers to cover two frequency bands 10 Hz-1 kHz, and 1 kHz-100 > kHz, and each has four outputs for the decade dividing. The switching > elements need to have as low an on resistance (<< 1 ohm) as possible, > and function well at extremely low levels. Since this is an AC system, > I'm not worried about Seebeck effects, just the low level contacting > ability. The primary sides of the transformers are at more normal > levels, so the switches there are not critical. > > I have all sorts of nice low level regular and Hg reed relays that would > do the job, but since the output side is to be isolated, to minimize > interference, it would be much better to do all the switching passively > with mechanical switches. This would avoid needing electricity, and the > proximity problems of having relay coils and capacitance in the low > level environment, and power supply and line noise and ground loops (no > power cord), or messing with battery power. Especially, there would be > no power transformer emissions to worry about. > > I have lots of mechanical switches of all sorts. I'd like to go with a > rotary wafer type to select the transformer taps and frequency ranges. I > have mostly standard Ag plated type contact ones and some Au ones. I can > usually build a switch from pieces for almost any arrangement. > > Another, but quite complicated option is to use low level reeds actuated > by a mechanically driven magnet, but doing that would be a big project > in itself, so scratch that, unless a simple way pops up. Another way is > to make heavy analog switches with big MOSFETs, but that's a lot of > parts, a battery, and issues with capacitance and crosstalk. So, good > old mechanical switching seems the best way to go - if only it can > actually be done. > > The big question then, is how to get a regular, environmentally exposed > contact to be usable in the nV region. Wetting the contacts with Deoxit > is the only practical thing I can picture so far. There is another class > of contact "protectors" based on synthetic oils, but I think I need > chemical action too, so these may not work - I'll have to investigate > that too. > > So anyway, does anyone have experience or knowledge of how Deoxit would > behave for very low level contacts? It may boil down to a "try it and > see" scenario, but it would be nice to have some info in advance." > > Now: > > I have opted for the "try it and see" scenario, and have built it up so > far using mechanical rotary wafer switches. The low-level section will > use all gold plated contacts, with Deoxit G100 applied. Also to be > included, is a "contact cleaning" mode, where some higher DC (couple > volts and mA) can be temporarily applied to all contacts from a built-in > Li cell circuit. Since I don't have enough of the right kinds of Au > plated contact wafer parts available, the higher level input (primary) > side will include some regular Ag plated contact switching too. These > will be chemically cleaned first, then have Deoxit D100 applied. > > I have it mostly mechanically built, almost ready to begin wiring up and > experimenting. This will be ongoing for a while, and I would appreciate > if anyone here does have experience with such low level applications > with exposed mechanical contacts. I know we usually talk about DC here, > while this is an AC application, but the same principles apply. > > Ed > > > > _______________________________________________ > volt-nuts mailing list -- volt-nuts@lists.febo.com > To unsubscribe, go to > http://lists.febo.com/mailman/listinfo/volt-nuts_lists.febo.com > and follow the instructions there. >
SW
Steve Wiseman
Mon, Jun 8, 2020 9:17 AM

If you want the pre-built goodness of reed switches, without the

electrical  interference from the coils and drivers, then how about
magnets-onna-stick as your switch control element? Either push-pull like an
organ stop, or rotate on a shaft? Easy to route to a front panel, cheap,
stable, mostly good things?

Ah, sorry, just seen you addressed this in your OP.  It still feels worth
investigating. I guess I just don't like oils long-term, in my experience
they go where they shouldn't, and won't stay where they should, and then
they go crusty / get dirty.
Still, do report back! At least it's AC and voltage, rather than charge...
How long does this thing have to last?

Steve

> If you want the pre-built goodness of reed switches, without the electrical interference from the coils and drivers, then how about magnets-onna-stick as your switch control element? Either push-pull like an organ stop, or rotate on a shaft? Easy to route to a front panel, cheap, stable, mostly good things? Ah, sorry, just seen you addressed this in your OP. It still feels worth investigating. I guess I just don't like oils long-term, in my experience they go where they shouldn't, and won't stay where they should, and then they go crusty / get dirty. Still, do report back! At least it's AC and voltage, rather than charge... How long does this thing have to last? Steve
DD
Dr. David Kirkby
Mon, Jun 8, 2020 9:32 AM

On Sun, 7 Jun 2020 at 14:55, Didier Juges shalimr9@gmail.com wrote:

It seems to me one of the problems you will fight with mechanical switches
is oxidation, or corrosion in a more general sense. It is hard
(impossible?) to make completely sealed mechanical switches.

I don't know if anything is commercially available, but I would have
thought a combination of

  • hermetically sealed enclosure for the switch
  • hermetically sealed contacts where the wires go in/out
  • switch mechanism operated by a magnetic field inside the hermetically
    sealed envelope
  • A mechanical switch, that moves a magnet outside the hermetically sealed
    envelope

would make it possible to create a hermetically sealed switch that is
mechanically operated from the outside. I would imagine such a device if it
exists would be quite expensive, and quite expensive to make, but I would
have thought it technically possible to make.

On Sun, 7 Jun 2020 at 14:55, Didier Juges <shalimr9@gmail.com> wrote: > It seems to me one of the problems you will fight with mechanical switches > is oxidation, or corrosion in a more general sense. It is hard > (impossible?) to make completely sealed mechanical switches. > > I don't know if anything is commercially available, but I would have thought a combination of * hermetically sealed enclosure for the switch * hermetically sealed contacts where the wires go in/out * switch mechanism operated by a magnetic field inside the hermetically sealed envelope * A mechanical switch, that moves a magnet outside the hermetically sealed envelope would make it possible to create a hermetically sealed switch that is mechanically operated from the outside. I would imagine such a device if it exists would be quite expensive, and quite expensive to make, but I would have thought it technically possible to make.