[time-nuts] Time Interval Algebra?
John Ackermann N8UR
jra at febo.com
Wed Dec 22 00:53:42 UTC 2004
Hi Bill --
You're right that to get a drift value you only need two measurements
spaced at the appropriate time interval for what you want to measure.
It gets more interesting when you want to look at stability more
closely. Attached is the phase plot I've been running for the last few
weeks. If you look at the first 17 days, you have a nice clean
accumulating offset of about -1x10e-12. But at day 18 something happens
(who knows what -- I was in India at the time, and there's no sign of a
temperature change or any other factor to cause a change) and the offset
flattens out. Then five days later it starts trending down again. If I
only measured from days 19 through 21, I'd think the oscillator had
virtually zero offset.
So, to see what the long term stability looks like, I'm taking readings
at regular intervals over, well, a long term. In this case, I'm using 1
second intervals because that's what the output is from the GPS, but
each line in the data file is an average of 300 readings (5 minutes)
because with anything shorter the GPS noise overrides any useful
information. Averaging more readings also increases the effective
resolution of the counter.
With a set of these equally spaced phase readings, I can do an Allan
Variance calculation that will quantify the stability of the oscillator
over various averaging times. I've also attached an AVAR plot of the
run so far. (By the way, the tool I use for this analysis is a
commercial package called Stable32, but as Tom noted you can generate
the AVAR values from a fairly simple piece of code.)
Noise in the divider chain can be a problem, but for these long-term
measurements it doesn't add too much uncertainty. The divider I'm
currently using is a simple set of 74LS390 chips, but I'm about to
replace that with a PIC-based single chip divider based on a design that
Tom did a few years ago. I've worked with a friend who has more PIC
skills than I to modify Tom's code to work with a 5MHz, rather than a
10MHz, clock.
John
----
Bill Hawkins wrote:
>Why is it important to get every measurement? If I turn the counter on
>Monday afternoon, get a reading of 0.500000 seconds, turn it off and
>next Monday afternoon get a reading of 0.500027, then I have an offset
>or error of 27 +/- 1 microseconds in 7 days. Each reading has been accurate
>to 1 microsecond, which requires that the counter clock drift less than
>0.1 microsecond per second.
>
>I suppose that the problem is that this doesn't allow an Allan Variance
>calculation based on one calculation per second. Why must it be one per
>second, not one in 10 or 100 seconds? Or why not 100 per second?
>
>A larger problem may be the variance in the counter that is dividing the
>Rb output into 1 pulse per second. Are the counters all clocked by the
>Rb clock or is it a simple cascade of decade counters?
>
>Bill Hawkins
>
>
>_______________________________________________
>time-nuts mailing list
>time-nuts at febo.com
>https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>
>
>
>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: hp5065a-gps.png
Type: image/png
Size: 6808 bytes
Desc: not available
URL: <http://febo.com/pipermail/time-nuts_lists.febo.com/attachments/20041221/c44de85a/attachment.png>
-------------- next part --------------
A non-text attachment was scrubbed...
Name: hp5065a-gps-sigma.png
Type: image/png
Size: 8953 bytes
Desc: not available
URL: <http://febo.com/pipermail/time-nuts_lists.febo.com/attachments/20041221/c44de85a/attachment-0001.png>
More information about the Time-nuts_lists.febo.com
mailing list