Thanks, Tom. That's another excellent point. The long term stability
impact of this cap is worth considering.
Following up on Brian's comment yesterday, I'm investigating sources for
very low leakage caps. If I find some, I'll let the group know, and if
there's a minimum purchase quantity that's reasonable, I may just buy a
bunch and make them available to the group here.
John
----
buehl wrote:
> John:
>
> If circuit has an electrolytic, following may influence how/ when you
> recalibrate.
>
> Electrolytics change value and leakage with no voltage applied for long
> periods. So if the long TC cap has been switched out for months it will
> need a long time with "in circuit" to stabilize again. The electrolytic
> barrier "forms" as a result of the electric field across it.
>
> Tom Buehl
>
>
> At 07:41 PM 4/5/2005, you wrote:
>
>> Thanks, Brian. That's very helpful. I looked at the schematic and
>> the cap they are using for LTC is 8uF that's simply put between the
>> input and the output of the op amp module.
>>
>> I guess the question is whether the leakage causes any stability
>> problems, or just a constant offset. Time for some experiments...
>>
>> John
>> ----
>>
>> wa1zms at att.net wrote:
>>
>>> John-
>>>
>>> I have seen the same effect on my HP-5060A. I had attribute it to
>>> leakage in the loop filter capacitor.
>>>
>>> Should you find that you need to replace that cap in your standard,
>>> you might want to try some of the very low leakage current caps from
>>> WIMA in Germany. If you're in a pinch I have a couple 1uF WIMA caps
>>> that I could send you.
>>>
>>> -Brian, WA1ZMS
>>>
>>> -----Original Message-----
>>> From: time-nuts-bounces at febo.com [mailto:time-nuts-bounces at febo.com]On
>>> Behalf Of John Ackermann N8UR
>>> Sent: Tuesday, April 05, 2005 9:53 AM
>>> To: Discussion of precise time and frequency measurement
>>> Subject: [time-nuts] 5061 Cs frequency shift with long time constant?
>>>
>>>
>>> This is very odd (I seem to be collecting these sorts of problems)
>>> and I wonder if anyone else has seen something similar.
>>>
>>> I got my 5061A trimmed to the point where a 5 day run against GPS
>>> showed an offset of less than 1x10e-13. I then switched the unit
>>> into the long time constant mode and now see that the offset (over 36
>>> hours) is -1.7x10e-12. Nothing else in the setup has changed.
>>>
>>> It's hard for me to see how simply changing the loop time constant
>>> would change the frequency of the standard, especially by this much,
>>> but stranger things have happened.
>>>
>>> Has anyone else experienced a shift like this when switching the time
>>> constant?
>>>
>>> Thanks,
>>>
>>> John
>>>
>>> PS -- I was able to partially, but not completely, reproduce the
>>> phase shift that I wrote about a couple of weeks ago by jiggling the
>>> 1 PPS cable to the counter. Replacing that cable seems to have
>>> resolved that problem.
>>>
>>> _______________________________________________
>>> time-nuts mailing list
>>> time-nuts at febo.com
>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>
>>> _______________________________________________
>>> time-nuts mailing list
>>> time-nuts at febo.com
>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>
>>>
>>>
>>
>>
>> _______________________________________________
>> time-nuts mailing list
>> time-nuts at febo.com
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>
> _______________________________________________
> time-nuts mailing list
> time-nuts at febo.com
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>
>
More information about the Time-nuts_lists.febo.com
mailing list