[time-nuts] Testing 2 Cesiums & GPS
John Ackermann N8UR
jra at febo.com
Wed Jan 19 14:51:51 UTC 2005
Brooke Clarke wrote:
> Something is very wrong with s/n 1013. But the green lock light is one.
> When I turn off the modulation and open the loop, sweeping the control
> voltage produces a definite peak. If the true peak was outside the
> control voltage range how much frequency error would there be?
>
Hi Brooke --
When Corby Dawson walked me through firing up my 5061A, he had me first
do a rough check with a frequency counter to look for 5.000 000 MHz
because if the C field was far enough off you could get a lock on a
secondary peak (there are seven total, with the two nearest the real
peak being the strongest) which would result in a frequency error of
about 0.25Hz . It looks like your error is much less than that, but it
might not hurt to double check and make sure you're on the right peak.
By the way -- I've been doing some runs of the 5061A versus Z3801A and
was quite pleased with the offset I apparently had -- after two days, it
was statistically flat (a few parts in 10e14). But then last night I
redid the rack mount which I'm sure jarred the unit a bit, adjusted the
loop gain as after several days of run time it had drifted up from the
nominal 40 to 50, and switched to the long time constant. I also
adjusted the trigger point on the counter. Now, a 12 hour run is
showing about -1x10e-12. Not sure which of the changes caused that, but
it's certainly clear to me now that this is a very touchy business!
I'm now dumping all the plots I generate (using the stable-stats.pl tool
I mentioned a week or so ago) into http://www.febo.com/time-freq/plots/
if you're interested. Not a lot there now, but I hope to soon be doing
some simultaneous intercomparisons of the Cs, Rb, and GPSDOs that'll be
plotted there.
John
More information about the Time-nuts_lists.febo.com
mailing list