[time-nuts] Some results of PRS10 and Trimble Resolution
Poul-Henning Kamp
phk at phk.freebsd.dk
Tue Jun 27 14:33:44 UTC 2006
In message <20060627134736.6784.qmail at web53702.mail.yahoo.com>, Radio Engineer
writes:
> I still don't really understand why the PRS10 Rubidium has such a bad 2s
>Allan Deviation with a 7-hour 1PPS filter time, also considering this kind of
>nice low Phase noise.
If you feed a PRS10 with a 1PPS that has systematic jitter (NB: not
random jitter!) like the Oncore GPS receivers have, then some heavy
duty averaging is called for before the loop is fed.
Unfortunately the 1/256'th exponential filter provided in the PRS10
is not long enough to fill this need.
Newer versions of the PRS10 firmware allows you to set the PPS input
offset, and with a small program you can transmit the negative sawtooth
correction from the GPS to the PRS and that solved the problem for me.
--
Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
phk at FreeBSD.ORG | TCP/IP since RFC 956
FreeBSD committer | BSD since 4.3-tahoe
Never attribute to malice what can adequately be explained by incompetence.
More information about the Time-nuts_lists.febo.com