Jack Hudler wrote:
> Sheeesh!
> Actually I use the 5501B for its intended purpose... I'm a bit of a precision
> nut. Never seen an Allan variance plot of one.
>
> I prefer 473.6057788309637 THz :)
>
> -----Original Message-----
> From: time-nuts-bounces at febo.com [mailto:time-nuts-bounces at febo.com] On Behalf
> Of Robert Atkinson
> Sent: Tuesday, November 07, 2006 2:55 AM
> To: Discussion of precise time and frequency measurement
> Subject: Re: [time-nuts] HP 10811A/B OCXO OP/SRV manual is onlineathparchive.com
> AND HP 5501B manual.
>
> Hi Jack,
>
> Great scan!
>
> I was also most interested to see you had scanned the manual for another
> HP reference that I have - the 5501B. With an accuracy of 1x10-7 and
> stability of 2x10-8 it's not that great but the frequency is 473612.23
> GHz! For those who have not caught on yet, the 5501B is a stabilised
> laser with about 1mW of 632.991372nm (red) output.
>
>
>
> Many thanks,
>
> Robert G8RPI.
>
>
>
>
>
> ________________________________
>
> From: time-nuts-bounces at febo.com [mailto:time-nuts-bounces at febo.com] On
> Behalf Of Jack Hudler
> Sent: 06 November 2006 19:31
> To: 'Discussion of precise time and frequency measurement'
> Subject: [time-nuts] HP 10811A/B OCXO OP/SRV manual is online
> athparchive.com
>
>
>
> I just finished and posted the 10811A/B OP/SRV to hparchives.com.
>
>
>
> http://www.hparchive.com/Manuals/HP-10811AB-Manual.pdf
>
>
>
> Enjoy,
>
> Jack
>
> --------------------------------------------------------
> Any opinions expressed in this email are those of the individual and not
> necessarily Genetix Ltd (Genetix) or any company associated with it. This email
> and any files transmitted with it are confidential and solely for the use of the
> intended recipient. If you are not the intended recipient or the person
> responsible for delivering to the intended recipient, be advised that you have
> received this email in error and that any use is strictly prohibited. If you
> have received this email in error please notify Genetix by telephone on +44
> (0)1425 624600.
>
>
> The unauthorised use, disclosure, copying or alteration of this message is
> strictly forbidden. This mail and any attachments have been scanned for viruses
> prior to leaving Genetix network. Genetix will not be liable for direct,
> special, indirect or consequential damages as a result of any virus being passed
> on, or arising from alteration of the contents of this message by a third party.
> _______________________________________________
> time-nuts mailing list
> time-nuts at febo.com
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>
>
>
>
> _______________________________________________
> time-nuts mailing list
> time-nuts at febo.com
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>
>
To measure the Allan variance you need at least 2 (preferably more)
stabilised lasers and a mixer (photodiode) plus a suitable amplifier to
produce a beat signal for analysis.
The beat frequency may be as high as 100 MHz with a pair of 5501s
(frequency/wavelength accuracy of 0.1ppm) so the photodiode and
associated amplifier will need adequate bandwidth.
Bruce