[time-nuts] How to measure Allan Deviation?
Dr Bruce Griffiths
bruce.griffiths at xtra.co.nz
Tue Oct 24 02:30:39 UTC 2006
Didier Juges wrote:
> Here are my $0.02...
>
> Magnus Danielson wrote:
>
>> From: Dr Bruce Griffiths <bruce.griffiths at xtra.co.nz>
>> Subject: Re: [time-nuts] How to measure Allan Deviation?
>> Date: Tue, 24 Oct 2006 10:52:21 +1300
>> Message-ID: <453D3995.8080004 at xtra.co.nz>
>>
>>
>>
>>> Tom
>>>
>>>
>> Hi Bruce,
>>
>>
>>
>>> In comparing 2 oscillators do you mean
>>>
>>> 1) Connecting one oscillator to the FREQ STD input at the rear of the
>>> 5370A, selecting the external timebase and connecting the other
>>> oscillator to the Front panel FREQ/PERIOD input and then selecting
>>> frequency measurement which in effect gives the frequency ratio of the 2
>>> oscillators?
>>>
>>>
>> I have advocated this approach, but nobody seemed to care.
>>
>>
>>
> HP says in the 5316B manual (probably in other manuals, but that's where
> I saw it first), that it is best to compare 2 oscillators by plugging
> them in START and STOP respectively, but NOT to use one as a standard,
> because averaging works better (or simply, works) when the internal
> standard is async with either START or STOP signals.
>
>> Cheers,
>> Magnus
>>
>>
>
> Didier
> _______________________________________________
> time-nuts mailing list
> time-nuts at febo.com
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>
>
This may also be helpful in mitigating the effect of some of the
differential linearity errors of the HP5370A
Bruce