[time-nuts] need recomendation for a portable 10mhz reference oscilator
Bruce Griffiths
bruce.griffiths at xtra.co.nz
Tue Dec 11 21:05:49 UTC 2007
Hal Murray wrote:
>> A good (and well-aged) double-oven OCXO with aging compensation should
>> be able to provide that type of performance. The problem is: how
>> much physical movement will the unit experience? What are the
>> temperature extremes that are expected, and how fast does the
>> ambient move between the temp extremes? Can it be GPS locked during
>> that time?
>>
>
> Assuming you don't GPS lock it...
>
> How well will two units from the same manufacturer/batch track?
>
> I'm assuming they will see roughly the same temperature profile and that you
> start with them running at the same frequency.
>
>
> Has anybody made a triple oven setup? I'm thinking of a big box with a few
> holes and a temperature controlled fan. The idea is to let the oven (and
> electronics) keep itself warm when the ambient temperature gets very cold.
>
> I'm picturing an open hill top at night with clear skies so there would be
> lots of radiation cooling. Of course, with an open hill top, GPS recption
> would be easy.
>
>
>
>
Hal,
A triple oven is a particularly bad idea, its just too difficult to
achieve a wide operating temperature range without losing temperature
control or cooking everything in the innermost oven unless one uses a
watercooled peltier device to control the outer oven temperature.
A well designed single oven with feedforward (or the outer oven of a
dual oven system if you cant redesign/rebuild the inner oven) can easily
achieve a thermal gain of more than 1E5 see:
http://tf.nist.gov/timefreq/general/pdf/595.pdf
A bootstrapped oven design like that used by Wenzel (not as new as he'd
have you believe - its been in use in one form or another for the better
part of a century) will also help in improving performance. The optimum
solution is of course to combine the 2 techniques.
http://www.wenzel.com/documents/Sub-pico%20Multiplier.pdf
With a thermal gain of 1E5 for the outer oven temperature fluctuations
at the outer shell of the inner oven will only be a 1 millikelvin or so
(assuming the ambient temperature range is less than 100K).
Bruce