> in my opinion, you would probably introduce some thermal noise through this
> 50KOhm equivalent resistor while the PLL is not operating. Noise could maybe
> be reduced if there is a cap to ground on this divider (this cap being part of
> the loop filter).
As long as you don't think it will affect stability when in lock, then I am ok.
> The PLL bandwidth depends solely on the phase noise of your 10MHz source
> versus your 100MHz oscillator. If the 10MHz source is better at say 100Hz offset
> (better by more than 20dB) then the loop bandwidth should be more than
> 100Hz, so that the PLL can actually reduce the phase noise of your 100MHz
> oscillator.
>
> At 100Hz offset, you say you have -68dBc/Hz at 100MHz. This calculates to
> -88dBc/Hz at 100Hz offset for the 10MHz source (excluding the PLL chip and
> loop-filter noise). You may want to check the noise performance of the PLL on the
> ADI website's PLL simulator.
>
> So if you have much better than -88dBc/Hz at 100Hz on your 10MHz oscillator
> (not hard to achieve, many oscillators have <-140dBc/Hz at 100Hz already)
> then you would be wasting performance with a <100Hz loop filter, and you may
> want to do a 1KHz or even wider loop filter or so. But if you don't know the
> 10MHz source's performance, it is probably best to be safe and use 10Hz, or
> 100Hz loop filter BW.
>
> A good spectrum analyzer (such as HP 8560B/E etc with the phase-noise
> software option) should allow you to measure <-68dBc/Hz noise at 100Hz offset at
> 100MHz, so you can check what BW results in the overall lowest noise.
>
> In short, if you want to maximize your systems performance, then loop
> bandwidth depends on the performance of the 10MHz versus the 100MHz oscillator for
> close-in phase noise.
>
> To do the math, subtract 20log(100/10) = -20dB from the noise of the 100MHz
> oscillator to get the equivalent noise energy for the 10MHz oscillator (at the
> same frequency offset).
>
> Hope this makes sense,
Yes, I understand all the math. I just don't know what to expect from
the 10 MHz references people will plug in. Clearly if they plug in a
Fury, I should use a wide bandwidth, but who knows what else they'll
use...
Thanks,
Matt