[time-nuts] HP 5370B question
Robert E. Martinson
REMartinson at rcn.com
Sat Dec 22 06:56:41 UTC 2007
I ran my 5370B for one hour, counting its own reference with a one second
gate. The summary of the results are:
min 9999999.99650 Hz
max 9999999.99963 Hz
avg 9999999.99796 Hz
median 9999999.99787 Hz
Per the manual's Performance Test 5 (page 4-12) the spec is +/- 0.005 Hz.
It would be nice if my average was 10,000,000.000 HZ but?? From a Time-nuts
thread of several months ago, the adjustment for this result is very touchy
and it's likely you will make the result worse rather then better. Since
mine is in spec, I'll leave it alone.
I have sent the graphical & raw data to Corby, if anybody else is interested
please advise me of your email address.
Regards,
Bob Martinson, N1VQR
-----Original Message-----
From: time-nuts-bounces at febo.com [mailto:time-nuts-bounces at febo.com] On
Behalf Of corby d dawson
Sent: Wednesday, December 19, 2007 2:15 PM
To: time-nuts at febo.com
Subject: [time-nuts] HP 5370B question
Hi,
Could a few of the 5370B owners connect the counters reference output to
its counting input jack and with a 1 second gate let me know what the
last 2 digits are doing?
An earlier post implied that the last couple digits will wander around
due to the way its designed.
I'd like to know what a couple different counters show to see if that is
indeed true!
Thanks,
Corby Dawson
_______________________________________________
time-nuts mailing list -- time-nuts at febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list