I ran into this same problem with my first SR620. I thought it was
defective, so I sent it to SRS for a cal and refurbishment. It got
better, but still had an offset of 0.0004 Hz. Later I bought a second
one, and it, too, had an offset similar to yours.
A careful reading of the manual, page 63, Performance Tests, Accuracy
(where the 10 MHz output from the back panel is measured at the A input)
reveals that a +/- 0.0035 Hz offset on a 10 MHz input is acceptable and
within spec. I think this also applies to your test configuration.
I have a photograph of my two SR620s, both using an external clock
(PRS-10/GPS from an FS710 Distribution Amp), and both measuring the same
signal on their A inputs. One reads 10,000,000.00096 Hz and the other
reads 9.999,999.99827 Hz. I used a gate time of 1 second averaged for
100 readings. Both SR620s had fresh factory calibrations.
Bob Crawford
TheInfamousFlavio at hotmail.com wrote:
>I have an SR620 counter that I set up with a Z3801A as an external reference. If I put a bnc T connector at the output of the Z3801A and use two equal length bnc cables, one to the ext. ref input on the back and the other to channel A then do a frequency measurement, I get a mean that is about .0015 Hz below 10,000,000.0000Hz.
>
>Does any know why this might be happening? I would expect it to read 10,000,000.0000 exactly give or take a couple on the last digit.
>
>_______________________________________________
>time-nuts mailing list
>time-nuts at febo.com
>https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>
>
>
>
More information about the Time-nuts_lists.febo.com
mailing list