[time-nuts] Fine print on HP 5334B
jshank
j-shank at comcast.net
Fri Feb 23 14:20:51 UTC 2007
Hi,
I check my 5370b by using the 10 MHz output and feeding it to the counter.
What I see is a fluctuation in the displayed reading of between 9.999 999
998 73 to 9.999 999 997 25. The instrument is set for 1 s gate. Sometimes
one reading will remain for a couple of arming but usually the reading will
change between six different readings with the high and low as above.
I am assuming there is a problem with this instrument and am asking for
suggestions on what can be done to correct this problem.
Ps. this instrument has the 10811-06111 oscillator. I have tried using the
10 MHz output from the 5370 as an external reference to a 53131A while
measuring the frequency of the 5370 10 MHz output. After both instruments
warm up the frequency of the 5370 output is displayed on the 53131A with a 1
s gate time as 10.000 000 000 and if I increase the gate time to 10s the
reading is 10.000 000 000 0 with the reading remaining rock solid.
Thanks,
Jeff
----- Original Message -----
From: "John Miles" <jmiles at pop.net>
To: "Discussion of precise time and frequency measurement"
<time-nuts at febo.com>
Sent: Thursday, February 22, 2007 3:29 PM
Subject: Re: [time-nuts] Fine print on HP 5334B
> What do people observe on their 5370A/Bs? After warmup, I tend to see a
> bias of about +0.0005 Hz, or 5E-11, when observing its own 10 MHz source.
> There is some occasional random waffling on the order of about +/- 0.001
> Hz
> but for the most part it reads close to 10.000 000 000 5.
>
> I have never been able to calibrate the interpolators (specifically the
> anticoincidence adjustment) in these counters, due to lack of a
> vernier-adjustable pulse generator. When I was going through one last
> week,
> I just left the anticoincidence pot in the middle of its range, and
> couldn't
> see any ill effects from doing so. The manual is not that great at
> describing exactly what this adjustment does.
>
> -- john, KE5FX
>
>
>> -----Original Message-----
>> From: time-nuts-bounces at febo.com [mailto:time-nuts-bounces at febo.com]On
>> Behalf Of Hal Murray
>> Sent: Thursday, February 22, 2007 12:10 PM
>> To: time-nuts at febo.com
>> Subject: [time-nuts] Fine print on HP 5334B
>>
>>
>> (Interesting timing given the SR620 discussion.)
>>
>> I'm (finally) getting my GPIB gizmo working.
>>
>> I have a HP 5334B running off its (good) internal osc looking at
>> the output
>> of a Z3801A. Assuming the Z3801A is stable, this is a backwards way of
>> looking at the internal osc. (There is a shared in/out connector for the
>> internal clock, so I can't see the internal osc if I feed it an ext
>> reference.)
>>
>> I left it running for a night with the mostly default setup and 100 cycle
>> averaging. I think the default gate time is 0.3 seconds. Answers ranged
>> from:
>> F +9.999999330E+06
>> F +9.999999333E+06
>>
>
>
> _______________________________________________
> time-nuts mailing list
> time-nuts at febo.com
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>
More information about the Time-nuts_lists.febo.com
mailing list