[time-nuts] Bad batch of HP10811's
Jeroen Bastemeijer
J.Bastemeijer at TUDelft.nl
Tue Jan 16 13:17:22 UTC 2007
Dear Time-nuts,
Just a quick experiment. I started with the cold 10811 OCXO (switched
off for >24 hours). I connected it to a counter and noted the frequency
every 30 seconds. (last two measurement each 5 minutes).
You can see the curve in the attached PDF. The offset at t=0 was 1706 Hz
(Compared to 10.000000MHz).
After 10 minutes warm-up the total offset is 0.75Hz (compared to
10.000000MHz) and offset is 0.12Hz from the "final" value. According to
the spec the offset should be within 5*10^-9.
From this I conclude the oscillator is not faulty. Sorry, Didier ;-)
The possible reason for this offset is probably not the crystal, but the
tempco of the rest of the circuit. The oven heats up till about 82
degrees centigrade. A PN-juntcion at room temperature (22 degrees) will
increase 60 degrees in temperrature, it's forward voltage will change by
approx. 120 mV!!!!! This change will lead to a significant change in
circuit properties (e.g. of the oscillator and the attached AGC circuit).
Probably this was the reason HP specfied it's oscillator first after 10
mins of warm-up.
This makes me wonder: If an oscillator gives a big offset, the
ovencontroller may be faulty... not working at all, or not getting warm
enough.
Jeroen PE1RGE
The raw data:
Time Frequncy [MHz] Offset [Hz]
0 9,998294 1706
30 9,998405 1595
60 9,998645 1355
90 9,998854 1146
120 9,999038 962
150 9,999197 803
180 9,999349 651
210 9,999466 534
240 9,999569 431
270 9,999666 334
300 9,99974 260
330 9,999804 196
360 9,999858 142
390 9,999906 94
420 9,99994027 59,73
450 9,99996696 33,04
480 9,99998633 13,67
510 9,99999979 0,21
540 10,00000062 -0,62
570 9,99999958 0,419999999
600 9,99999925 0,75
900 9,99999933 0,67
1200 9,99999937 0,630000001
Didier Juges wrote:
>That sounds like an awful lot of drift, absolutely inconsistent with a
>crystal oscillator of any kind.
>
>Either this OCXO is really sick, and just happen to end up at the right
>frequency by accident, or it is an OCO (Oven Controlled Oscillator :-)
>
>Didier KO4BB
>
>Jeroen Bastemeijer wrote:
>
>
>>Dear Rick,
>>
>>Are you sure the 500Hz is too much for a cold oven? I checked my 5345A
>>(recently acquired) counter, which was switched off for some time.
>>Measuring the frequency directly after applying power and switching on,
>>resulted in about 1640 Hz frequency offset. After warm up, the error is
>>below 0.4 Hz. (The unit used here is a 10811a). The error drops pretty
>>fast, allthough I didn't measure an accurate curve. Is the frequency
>>difference caused by the tempco of the oscillator circuit?
>>
>>Cheers, 73s Jeroen PE1RGE
>>
>>
>>
>
>
>_______________________________________________
>time-nuts mailing list
>time-nuts at febo.com
>https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>
>
>
>
>
--
Ing. Jeroen Bastemeijer
Delft University of Technology
Department of Electrical Engineering
Electronic Instrumentation Laboratory
Mekelweg 4, Room 13.090
2628 CD Delft
The Netherlands
Phone: +31.15.27.86542
Fax: +31.15.27.85755
E-mail: J.Bastemeijer at TUDelft.nl
-------------- next part --------------
A non-text attachment was scrubbed...
Name: 10811coldstart.pdf
Type: application/pdf
Size: 10085 bytes
Desc: not available
URL: <http://febo.com/pipermail/time-nuts_lists.febo.com/attachments/20070116/c6aec5b2/attachment.pdf>
More information about the Time-nuts_lists.febo.com
mailing list