Brooke Clarke wrote:
> ); SAEximRunCond expanded to false
> Errors-To: time-nuts-bounces+bruce.griffiths=xtra.co.nz+bruce.griffiths=xtra.co.nz at febo.com
>
> Hi Didier:
>
> Would you elaborate on the comment "Gold plated connectors are a well known
> example." Do you mean when soldered with Lead Tin solder instead of a silver
> bearing solder or something else?
>
> Have Fun,
>
> Brooke Clarke
> http://www.PRC68.com
> http://www.precisionclock.com
>
>
>
> Didier Juges wrote:
>
>> ); SAEximRunCond expanded to false
>> Errors-To: time-nuts-bounces+brooke=pacific.net+brooke=pacific.net at febo.com
>>
>> I have seen cold solder joints on thermal fuses and certain types of
>> capacitors, while the rest of the instrument was fine with no sign of
>> corrosion.
>>
>> I think it has to do with the metal used for certain component leads. Either
>> they were never soldered well, or interface corrosion developed over time.
>>
>> Gold plated connectors are a well known example.
>>
>> Didier KO4BB
>>
Brooke
Gold dissolves in the solder and a gold -tin intermetallic compound is
formed which severely reduces the joint integrity and ductility.
This can be circumvented by keeping the gold concentration in the solder
below 4%. Tin plating over the gold before soldering is sometimes used
to ensure this.
Bruce