[time-nuts] another Ebay mixup, 5370
Dr Bruce Griffiths
bruce.griffiths at xtra.co.nz
Sun Jun 10 14:04:47 UTC 2007
Tom Van Baak wrote:
>> Measurements taken from 3325B at 1 MHz:
>>
>> 5370B reference: INT EXT (same 5087A as 3325B uses)
>> Square wave, 2v pk: 15-20 ps 11-12 ps
>> Sine wave, 2v pk: 50-70 ps 30-35 ps
>> Square wave, 1v pk: 20-25 ps 12-13 ps
>> Sine wave, 1v pk: 60-80 ps 40-45 ps
>> Square wave, .1v pk: 40-60 ps 40-45 ps
>> Sine wave, .1v pk: 600-900 ps 500-600 ps
>>
>> -- john, KE5FX
>>
>
> Very nice data. Thanks for staying up late to make the measurements.
> What trigger levels and input configuration did you use for this?
>
> Based on these results can someone calculate how much jitter is
> due to the 5370B input circuits alone and how much is due to the
> rest of the circuitry of the unit?
>
> /tvb
>
>
>
Tom
Since, at various times the HP5370 used a discrete ECL gate crystal
oscillator, an HP10544A, and an HP10811A as the internal reference, it
would also be useful to indicate which of these a 5370 uses as its
internal reference.
The time interval mode (+TI only or +-TI) should also be stated togeter
with input impedance settings and attenuator settings.
Taking the external 1MHz sinewave (EXT reference for HP5370) and
assuming input attenuator set to 1x, and the 2 inputs are connected in
parallel to the START BNC then the corresponding trigger circuit noise
for 0.05V pk input (input signal is attenuated by 2 at the trigger
circuit inputs for 50 ohm Zin ) is about 123uV rms for both the START
and STOP channels which seems plausible.This gives about 50ps noise
induced jitter at 1V peak amplitude. However since the time interval
mode isnt specified nor are the number of samples taken, these
calculations are somewhat uncertain.
Also the intrinsic jitter of the source is not known (to me at least).
At 2V peak the noise induced trigger jitter could be 25ps with other
noise say 10-12ps.
The SR620 has a wider input trigger circuit bandwidth and can thus be
expected to be somewhat noisier. Both counters use ECL and/or similar
performance logic for the high speed sections.
Bruce