[time-nuts] 5370B improvement report
SAIDJACK at aol.com
SAIDJACK at aol.com
Sat Jun 23 09:03:21 UTC 2007
Hi guys,
I am back from a week long Asia trip. Here are some results of my 5370B
modification/calibration running on the Jackson-Labs Fury 10MHz against a PRS10
for the entire week.
After noticing that the Frequency Multiplier board (10MHz to 200MHz)
trim-caps settings are really key to the RMS noise performance of the 5370B, I came
up with some modifications to the circuit to improve it's performance.
The goal of this was to get a 200MHz ECL signal out of the A21 reference
board with the least jitter:
1) take the 50MHz output of the board (TP2 on the A21 board for example) and
feed it into an ultra-low-noise crystal-oscillator 100MHz PLL. Set the PLL
loop-bandwidth very low, less than 1Hz to filter out all reference noise.
2) take the 100MHz crystal output, and feed it into Q1 to multiply it to
200Mhz
3) add a fast buffer before U1 to generate faster edge-rates into the ECL
comparator
The result of this works very well. The advantages are: the 50MHz tanks
don't need to be adjusted as well as they needed to be adjusted before, since the
PLL filters out the 50MHz jitter. The 200MHz tank adjustment is not as
critical as before since most of it is bypassed - only a single 200MHz trim-cap
remains to be tweaked.
I used a Crystek PN CVHD-950X-100.000 ultra-low-phase-noise crystal
oscillator (VCXO) for this, simply because it was available at Mouser.com and because
I could not get a 200MHz version.
It's 100MHz output is fed into a 74LVC74 divider to generate 50MHz, which in
turn feeds a 74LVC86 Exor used as a phase-comparator. The other side of the
phase comparator is fed by the 50MHz from TP2 (a buffered CMOS version of
it). This locks the very low jitter 100MHz Crystal output to the 10MHz 10811
OCXO reference.
The result is quite good, no more need for tweaking all the trim-caps!
The crystal certainly generates a more stable 100MHz signal than the
multiplier does, and it is much easier to get 200MHz from the 100MHz CMOS output
than from the 50MHz sine wave in the original circuit.
Of course an ECL/PECL/LVDS 200MHz VCXO would have done an even better job,
feeding it's output directly into U1 without any transistor tanks etc.
In the end, this is all that the unit cares about: a very stable low-jitter
200MHz ECL reference.
BTW: this 200MHz doesen't even need to be phase locked with the 10MHz
reference, I tried using a Jackson-Labs FireFox signal generator set to 200MHz and
running totally asynchronously and the resulting performance of the unit was
just as good.
a disclaimer: please note that I am not suggesting to anyone to modify their
5370B, this is done at your own risk.
The attached image is the Fury GPSDO running against my FTS-4050 Cesium for
some hours. The vertical scale is 14.5ns pk to pk. You can see the very low
inherent jitter of <30ps.
Bye,
Said
************************************** See what's free at http://www.aol.com.
-------------- next part --------------
A non-text attachment was scrubbed...
Name: 5370b_noise.pdf
Type: application/pdf
Size: 47713 bytes
Desc: not available
URL: <http://febo.com/pipermail/time-nuts_lists.febo.com/attachments/20070623/421a6027/attachment.pdf>
More information about the Time-nuts_lists.febo.com
mailing list