On Tue, 27 Nov 2007, Prologix wrote:

> Hello Chuck,
>
> It appears that 3478A sends data continuously as soon as it is addressed to
> talk. Since the Prologix adapter is busy processing GPIB data it is unable
> to respond to USB commands.
>
> One solution is to turn read-after-write off (++auto 0) before connecting
> the adapter to 3478A, and then use ++read command to read one measurement at
> a time. Please see the manual (www.prologix.biz) for ++read command options.
>
> The other option is to set 3478A to one-reading-per-trigger mode using "T3"
> device command. See HP3478 manual (pages 39, 59). Then use the ++trg command
> to trigger the device. ++auto must be set to 1 in this case.
>
> Regards,
> Abdul


Excellent answer!  I do not have either piece of equipment but I can certainly
say that even though I am now using NI GPIB cards which are likely faster, I
always program the instrument being queried to only send one response per
trigger.

I learned BASIC and also HPIB programming on a HP85 when it was the latest and
greatest in 1980-81. Then later in another company I used the 9845 and then
after that used the 9836. I think I must have learned from their wonderful
documentation sets that this was the right way to take readings. It may well
have been to avoid the timing issues you presented.