Also, if it's of any use, the HP-85 controller does an unlisten, 
my-talk-addr, listen-address-x for each read. There's no mention of the 
HP-85 doing an automatic group-execute-trigger, and none of the examples 
issue an explicit trigger.

There is one sentence in the 5334 manual that says a g-e-t will trigger 
a new measurement, which is what I would expect.

Bill Ezell
----------
They said 'Windows or better'
so I used Linux.



wje wrote:
>    That does the trick, thanks. Interestingly, the HP manual doesn't
>    discuss group-execute-trigger at all. It's possible that the HP-85
>    programmable controllers would do a trigger before a read; all of the
>    examples in the 5334A manual are written for that controller.
> Bill Ezell
> ----------
> They said 'Windows or better'
> so I used Linux.
>
>    Prologix wrote:
>
> You could try explicitly triggering with ++trg followed by ++read 10.
>
>
> -----Original Message-----
> From: [1]time-nuts-bounces at febo.com [[2]mailto:time-nuts-bounces at febo.com] On
> Behalf Of wje
> Sent: Wednesday, August 27, 2008 4:14 PM
> To: Discussion of precise time and frequency measurement
> Subject: Re: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?
>
> YI don't use read-after-write, and I explicitly set it off before
> reading. However, I did try it once just for fun, and it does work as
> expected.
>
> The 5334 doesn't seem to assert eoi, so I do a '++read 10', which works
> just fine except for the WA1 strangeness.
>
> Bill Ezell
> ----------
> They said 'Windows or better'
> so I used Linux.
>
>
>
> Prologix wrote:
>
> Bill,
>
> How are you reading from 5334A? Is the read-after-write mode OFF?
> I recommend the following sequence:
>
> ++auto 0        -- turn off read-after-write
> ++read eoi      -- read until EOI is asserted by the instrument
> ++read eoi
> ....
>
> (If your instrument does not assert EOI, specify a character to terminate
> ++read command. See manual for details.)
>
> Yes, ++read will address the instrument to talk, read any output, and then
> address it to listen.
>
> We are testing an update that resets the controller if it detects a hung
> GPIB bus. I can send that to you offline.
>
> Regards,
> Abdul
>
>
>
> -----Original Message-----
> From: [3]time-nuts-bounces at febo.com [[4]mailto:time-nuts-bounces at febo.com] On
> Behalf Of wje
> Sent: Wednesday, August 27, 2008 3:12 PM
> To: Discussion of precise time and frequency measurement
> Subject: [time-nuts] Prologix GPIB-NET, HP 5370A, HP5334A problem?
>
> I just got one of the new networked GPIB controllers, and I've been
> having some issues. I'm not sure if it's the Prologix or the
> instruments, or both.
>
> With the 5370A, I can get samples for some period of time ranging from
> about 2 to 4 hours before the GPIB controller stops responding. I
> suspect this one might be the 5370A -488 controller bug I've heard of.
> However, there is no way to recover without power-cycling the Prologix.
> Even if this is the 5370A, the Prologix should really have an
> asynchronous reset capability.
>
> With the 5334A, I can get samples for as long as I want, no problem.
> However, if I set the 5334A to 'WA1' mode, which is supposed to wait
> until the counter is addressed before taking each sample, what happens
> is that I can read one value. Every successive read returns the same
> value; the 5334A never triggers for a new sample. Either I don't
> understand the behavior of WA1, although HP's sample program indicates
> that what I've described is proper, or there is some incompatibility
> between the Prologix and the 5334A.
>
> Unfortunately, the documentation for the Prologix is scanty at best.
> There is no description of the actual IEEE-488 handshake process that's
> going on. For example, does the Prologix do an unlisten after a read?
> The HP docs seem to indicate that in WA1 mode, a sample starts when the
> 5334A is addressed. So, if it's never unlistened, perhaps it never takes
> a new sample?
>
> Any thoughts or suggestions would be appreciated.
>
>
>
> _______________________________________________
> time-nuts mailing list -- [5]time-nuts at febo.com
> To unsubscribe, go to
> [6]https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
> _______________________________________________
> time-nuts mailing list -- [7]time-nuts at febo.com
> To unsubscribe, go to [8]https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
> References
>
>    1. mailto:time-nuts-bounces at febo.com
>    2. mailto:time-nuts-bounces at febo.com
>    3. mailto:time-nuts-bounces at febo.com
>    4. mailto:time-nuts-bounces at febo.com
>    5. mailto:time-nuts at febo.com
>    6. https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>    7. mailto:time-nuts at febo.com
>    8. https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
>   




More information about the Time-nuts_lists.febo.com mailing list