[time-nuts] 35601A as stand-in for 11848A ?
Bruce Griffiths
bruce.griffiths at xtra.co.nz
Mon Jan 7 03:59:51 UTC 2008
SAIDJACK at aol.com wrote:
> In a message dated 1/6/2008 15:01:25 Pacific Standard Time, jmiles at pop.net
> writes:
>
>
>> 85671A package or mine, how _exactly_ are you turning the dBc/Hz values you
>> get from (say) the 8561E into your ~-120 dBc/Hz numbers? Let's hear the
>> step-by-step procedure on that, and the problem should reveal itself.
>>
> Hi John,
>
> by selecting a 1Hz RBW on the 8561E. The APII analyzer should have a 1Hz BW
> normalization as well, although I checked in the user manual again and it
> just specifies "dbm" without regards to RBW. How does one calculate RBW from an
> FFT? Is it the width of a given FFT bin? I will check this again, and try to
> increase the number of samples.
>
> But even with say a 10Hz RBW I would expect the noisefloor to only rise
> 10xlog(1/BW) or 10dB, much less error than what I am seeing.
>
> bye,
> Said
>
>
Said
1 Hz resolution bandwidth isnt the same as 1Hz noise bandwidth.
The noise bandwidth of of a 1Hz resolution bandwidth depends on the
details of the filter.
For an FFT analyser, selection of a Hann, Blackman-Harris, uniform or
other window function/filter changes the equivalent noise bandwidth of
the filter.
Bruce
More information about the Time-nuts_lists.febo.com
mailing list