Martyn wrote:
>
>I did do a plot selecting Phase Modulation Spectral Density (dB/Hz)
>on the graph instead of dBc/Hz. Here the spurs were 3 dB lower!!!
>
To make sure I understand your results, here's how I would
summarize your measurements, inserting some made-up numbers
for observed spur and noise floor levels:
RF analyzer directly measuring 10 MHz signal :
- spur at 10.000100 Hz is @ -50 dBc
3048A L(f) measuring coherently demodulated phase:
- spur at 100 hz is @ -56 dBc
- noise floor is -160 dBc/Hz at 5 MHz offset
3048A Sphi(f) measuring coherently demodulated phase:
- spur at 100 hz is @ -53 dB
- noise floor is -157 dB/Hz at 5 MHz offset
If this is what you are seeing for relative changes in level
for the different measurements, I think the 3048A is working
just fine.
Here's another experiment to try [1]:
- change spur to an offset of 500 kHz @ -50 dBm
this moves the spur into broadband analyzer territory
- re-run your L(f) and S(f) plots at this offset
I would expect the same 6 dB, 3dB spur results for L(f), S(f)
- change the measurement setup BW to zero Hz instead of 1 Hz,
and measure Sphi(f) again
This forces the 3048A to NOT apply bandwidth corrections
to the collected data, so it will display a segmented-looking
trace with increasing floor offset across each band break
Brian
[1] It's been about 7 years since I've actually pushed the buttons
on a 3048A, so my recollection of setup and S/W behavior might
be a bit fuzzy