The HP 5334A or B with the C channel option has a sensitivity spec of -30dBm
(from memory) up to 1 GHz or so. The C channel has 15mV rms sensitivity at 1
GHz.

The C channel option is rare, but it only requires 3 parts (a diviser and
two dual-shottkys) and a connector (and a hole in the front panel) to add it
to a 5334B for instance. I have two 5334Bs (one is broken) and both have the
other parts required for the C channel except for these three. There is a
socket for the diviser, the shottkys have to be soldered in.

The C channel input is activated on pressing the '9' key.

Unfortunately, the MB506 diviser (Fujitsu) seems a little hard to find. If
anyone has a spare, let me know. A surface mount equivalent should not be
too hard to put in.

The HP 5316 has similar characteristics and also has a 1 GHz option, which
is also rare.

I also have an Advantest TR5823 counter which has the 1.3 GHz input with
20mV rms sensitivity at 1.3 GHz.

The HP models have reciprocal counting, so they work MUCH better for low
frequency signals.

Didier KO4BB

> -----Original Message-----
> From: time-nuts-bounces at febo.com 
> [mailto:time-nuts-bounces at febo.com] On Behalf Of 
> Jean-Christophe Deschamps
> Sent: Saturday, March 08, 2008 10:49 AM
> To: time-nuts at febo.com
> Subject: [time-nuts] Which HP Frequency Counter?
> 
> Dear group,
> 
> I consider buying a used lab frequency counter / timer --preferably
> HP/Agilent-- covering from few mHz to perhaps 400 MHz.  I'm 
> worried that models handling high frequencies seem to be 
> limited to under 100 mV input signal max. I don't want to 
> destroy an input channel each other day when 
> calibrating/repairing some instrument.  Also are there 
> instruments with high impedance input? 50 Ohm is not quite 
> right for investigation in the guts of most designs.
> 
> 
> Is it possible to find a not-too-old model under $800?  I 
> would like to find something in France or UK or Europe, but 
> it seems hopeless in this budget.  I would favor a repairable 
> model (with available service docs & schematics).
> 
> Your expert advises are more than welcome!
> 
> --
> <mailto:jcd at q-e-d.org>jcd at q-e-d.org  
> 
> 
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com To unsubscribe, 
> go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
> 
> No virus found in this incoming message.
> Checked by AVG. 
> Version: 7.5.518 / Virus Database: 269.21.7/1319 - Release 
> Date: 3/8/2008 10:14 AM
>  
> 

No virus found in this outgoing message.
Checked by AVG. 
Version: 7.5.518 / Virus Database: 269.21.7/1319 - Release Date: 3/8/2008
10:14 AM
 





More information about the Time-nuts_lists.febo.com mailing list