At 09:40 PM 5/10/2008, Didier Juges wrote...
>In my real world ... there is no allowance for measurement or
>calibration uncertainty except in some very unusual circumstances. The
>customer reviews and approves the test procedure and acceptance
>criteria...
If the example you gave followed those rules, then there would have
been no need to mention the 1 dB accuracy of microwave network
analyzers.
Even by the rules you now introduce, if the analyzer measures 2.01 and
the spec says 2 max, there's a problem, but still you argue with your
customer that 2.01 is as good as 1.99. If the criteria are agreed upon
and that criteria says there is to be a maximum measurement of 2, why
do you argue?
It makes as much sense as bringing an example where measurement
uncertainty is ignored into a discussion of accuracy, precision and
resolution.
More information about the Time-nuts_lists.febo.com