[time-nuts] any way to bootstrap a frequency standard into a voltage or resistance standard?
Didier
didier at cox.net
Fri Nov 28 14:57:55 UTC 2008
Scott,
A relatively inexpensive voltage reference is to actually buy an HP 3456A
DVM. It is rated as a transfer standard, that means you send it to NIST to
have it calibrated, and then you can use it to calibrate other instruments.
I have two here and together they cost me $130.
It is also a 6 1/2 digit DVM with GPIB interface. The reference voltage is
not provided externally, but you would use the meter to check and correct an
external reference or stable voltage source.
The service manual is on my web site http://www.ko4bb.com/cgi-bin/manuals.pl
Didier KO4BB
> -----Original Message-----
> From: time-nuts-bounces at febo.com
> [mailto:time-nuts-bounces at febo.com] On Behalf Of Scott Burris
> Sent: Thursday, November 27, 2008 11:18 PM
> To: Discussion of precise time and frequency measurement
> Subject: [time-nuts] any way to bootstrap a frequency
> standard into a voltage or resistance standard?
>
> Now that many of us have a nice 10Mhz reference courtesy of
> TAPR, I was wondering if there was any way to use that to
> build a precise voltage or resistance standard?
>
> I've got once of those high precision standard resistors with
> a sticker on it noting the actual measured resistance. Is it
> still accurate? Who knows?
>
> As well, I have a +5v reference that uses an Analog Devices
> precision reference chip as its source. I have more faith
> that this reference is correct within the tolerances
> specified in the datasheet.
>
> Now if I could somehow take that frequency reference and
> derive a voltage standard or the like, I'd be in business.
> But I can't think of a way that wouldn't require calibration
> of some sort, and if I had the means to calibrate, I wouldn't
> need the standard in the first place.
>
> Any voltage-nuts or resistance-nuts out there?
>
> Scott
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com To unsubscribe,
> go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list