I can confirm that the Wavecrest is sensitive to edge rise-time. It was not
designed to measure sine waves.
With a 10MHz sine wave from a Fury GPSDO as the source I get 8 - 10ps rms
jitter.
That exact same signal run through an NC7SZ04 buffer prior to feeding it
into the DTS-2075, then scaled to the +/-1.1Vpeak input of the DTS-2075 yields
jitter below the units' noise floor (~3.2ps rms) on a good OCXO.
I also have a DTS-2070C that performs slightly better - about 2.9ps rms.
bye,
Said
In a message dated 4/4/2009 17:43:23 Pacific Daylight Time,
peterawson at earthlink.net writes:
Bruce,
Your analysis conforms closely to measured results on my DTS-2075.
With the cleanest 10MHz source I have, at 2Vp-p, the DTS-2075 jitter
reading is 11.4ps rms. Running this number back into equivalent input
noise yields 716uV rms. The DTS-2075 input spec (assumed to be
for -3dB response) is 700MHz, which should be nearly 900MHz
equivalent noise bandwidth. These numbers are all reasonable &
serve to demonstrate the need for low noise signal conditioning.
Pete Rawson