[time-nuts] Another 3458A problem
Don @ True-Cal
true-cal at swbell.net
Tue Aug 4 14:32:38 UTC 2009
Noticed I misquoted the ERRSTR. It is:
114 System Error – multislope rundown convergence
Regards...
Don
________________________________
From: "Don @ True-Cal" <true-cal at swbell.net>
To: Discussion of precise time and frequency measurement <time-nuts at febo.com>
Sent: Tuesday, August 4, 2009 9:29:15 AM
Subject: [time-nuts] Another 3458A problem
Hello,
Seems this is the month for errant 3458A DMMs. I have a unit that was working fine and passing both the power on and full Self Tests. The unit has been stored away for a few years and is very clean. After several days of continuous error free operation, is started kicking out the error:
114 System Error – multislope rundown error
At first, the error was only every few minutes (all default power on settings) but is now almost constant. I can clear the error and proceed for a few more seconds. It is also the only error in the stack.
Suspicious of electrolytics, I checked all voltages and ripple on both outguard and inguard supplies with everything looking good. I did replaced some caps that showed barely low values. This unit is 11 years old. At this point, I am reasonably sure it is not power supply related. I am also sure it is not a cal related condition as I started a procedure and it would not get past the first offset section – same error.
Does anyone have any documentation of knowledge of this error? ERRSTR 114 is not mentioned in the Assembly Level Repair manual. Thanks.
Regards...
Don
________________________________
From: Dick Moore <richiem at hughes.net>
To: time-nuts at febo.com
Sent: Sunday, August 2, 2009 2:08:08 PM
Subject: Re: [time-nuts] new request for 3458A info
On Aug 2, 2009, at 5:00 AM, time-nuts-request at febo.com wrote:
>
> Message: 4
> Date: Sun, 02 Aug 2009 03:00:06 +0200
> From: "Dr. Frank Stellmach" <drfrank.stellmach at freenet.de>
> Subject: [time-nuts] 3458A info
> To: time-nuts at febo.com
> Message-ID: <4A74E516.5020202 at freenet.de>
> Content-Type: text/plain; charset=ISO-8859-15; format=flowed
>
> Dick,
>
> did you already try an ACAL DCV, or ACAL ALL from the front panel?
>
> Maybe the false cal constants will be set correctly.
>
> In the calibration manual, its explained, which constants are set by
> atocal (and cured), and which one are set by basic calibration,
> requiring external standards (10 volt, 10kohm).
> I think, DAC Vos and 10V gain are set by ACAL.
>
> Frank
Yes, tried all that. Last night I gave up and opened the box, which I had been reluctant to do -- DC board had burned components right in the center. There are 2 unrecognizable ICs and a handful of resistors, etc, and a badly charred board. Probably unrepairable. Have no idea what could have happened to make it so bad without also killing the power supplies, which are AOK. Shouldn't have been anything on the input, which is protected against over-voltage/-current. Too bad.
Thanks for all the advice and tips.
Dick Moore
_______________________________________________
time-nuts mailing list -- time-nuts at febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
_______________________________________________
time-nuts mailing list -- time-nuts at febo.com
To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list