At 08:30 PM 12/1/2009, J. Forster wrote...
>In fact, on the Tek 7704 or 7704A (I forget which) there were "No
>Cost"
>options of maximally flat frequency response OR best pulse response.
>OR!
Sure, if you're using a 150 MHz to try and measure accurately at 100
MHz, you're not going to get absolute accuracy. I'm not familiar with
that particular scope, but would bet the difference appears as peaking
at the extreme. In the absence of circuitry which deliberately changes
it, the response of an analog scope is (very nearly) Gaussian.
Can you quantify the difference between those options when measuring a
50 MHz (1/3 BW) signal?
Since the OP is a ham, I assume his quest is relative to commonly
available power meters for that market, most of which can't be expected
to do better than 20% accuracy, maybe 5% at the high end (
http://www.telepostinc.com/lp100.html ).
Sure, you can buy an uncalibrated 3400A (good to ~150W), and get some
unknown amount worse than 5% accuracy. Or a 432a (good to a whole 10
mW!) for <$100, plus another $100+ for the mount/cable. Now you've got
something which is good to a couple of %, with an _extremely_ limited
range (and practically useless for many ham applications).
A Gaussian scope is predictably off ~3% @ BW*0.3. A decent scope will
have a vertical amp accurate to a couple of percent. With a 1% load,
you can measure at the 5% level of a $400 dedicated wattmeter (from < 1
uW to 400W, and to the 2000W ham limit if you invest in a 100:1 probe),
for little cost.
"Accurate...simple and inexpensive." Many non-appliance-operator hams
will already have a scope, so the cost is a $10 resistor (for a 100W
rated Caddock, which should be good for a few seconds of full power,
enough to make the measurement).
Finally, as I've already mentioned, one can make a simple peak detector
using a rectifier, and measure the DC voltage off that.