[time-nuts] 5370A/B measurement floors (was Re: Fixing an HP 5370A...)
John Miles
jmiles at pop.net
Sat Jul 25 16:00:50 UTC 2009
It's worth noting that the 5370's accuracy is supposedly never better than 1
ns (specifically, +/- resolution +/- timebase*interval +/- trigger jitter
+/- 1 ns systematic). We tend to rely on it to do much better than that --
or at least we all count on the "1 ns systematic" to look like random noise
in the long run -- but it still bears remembering, because that systematic
error swamps any other effects in the short term.
Some time ago John A. plotted his 5370's residual floor, about halfway down
the page at http://www.febo.com/pages/hp5370b/ . He fed a 1-pps signal into
a T-connector at the START jack, followed by 35 ns of coax to the STOP jack.
I ran a similar test recently, but with a TADD-2 1-pps divider driving the
START jack and a similar amount of coax driving the STOP jack from the 10
MHz input side of the TADD-2, trying to get an idea of the composite floor
for a typical measurement using the TADD-2, the 5370B, and a bunch of coax.
The 5370B was driven by a Thunderbolt for this test.
My results (attached) were a bit worse than John's, but with a nearly
identical slope for the duration of his test. I left it running for a few
days, and saw some significant flattening on timescales of 3-4 hours that
matched my air conditioning cycle time (as reported by the shop thermometer,
aka Lady Heather.) Based on this graph I don't consider temperature
variations to be a grave concern above floors of 1E-14, or conversely at
timescales less than an hour or two, but they should definitely be minimized
or eliminated for HP 5370A/B measurements below that.
I still need to try variations of this test with the internal reference,
with and without the TADD-2 between the START and STOP jacks, and with
10-second and 100-second means on the 5370B. Clearly it would be good to
achieve better than 1E-10 accuracy at t=1s, since most good crystal
oscillators are better than that.
-- john, KE5FX
> -----Original Message-----
> From: time-nuts-bounces at febo.com [mailto:time-nuts-bounces at febo.com]On
> Behalf Of Bruce Griffiths
> Sent: Saturday, July 25, 2009 6:18 AM
> To: Discussion of precise time and frequency measurement
> Subject: Re: [time-nuts] Fixing an HP 5370A with a missing ROM board
>
Such a frequency shift corresponds to about 2.5ns error in a 10s time
interval (5370 maximum range).
This is much greater than the 25ps resolution and 100ps maximum
differential nonlinearity spec for the 5370.
For time intervals greater than 100ms or so such frequency changes will
significantly degrade the performance.
Bruce
_______________________________________________
time-nuts mailing list -- time-nuts at febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
-------------- next part --------------
A non-text attachment was scrubbed...
Name: resid_floor.gif
Type: image/gif
Size: 22399 bytes
Desc: not available
URL: <http://febo.com/pipermail/time-nuts_lists.febo.com/attachments/20090725/094193e6/attachment.gif>
More information about the Time-nuts_lists.febo.com
mailing list