[time-nuts] HP 3458A ISOLATOR DTACK err @ pwr up
Brooke Clarke
brooke at pacific.net
Thu Jul 30 00:08:38 UTC 2009
Hi Ed:
I've fixed the schematic.
Yes, I plan to check the transformer and A4 board prior to connecting back to
the rest of the DVM.
Have Fun,
Brooke Clarke
http://www.prc68.com
Ed Palmer wrote:
> Hi Brooke,
>
> Three thoughts:
>
> 1. If CR11 or CR12 blew open, the LM317 and LM337 might have been
> damaged. Are your +18 & -18 supplies okay? By the way, I think you've
> got them reversed on the schematic. The LM317 should be connected to
> CR01 and CR02.
>
> 2. On the web page you mentioned that CR10 survived. Does that mean
> that CR09 blew open? If so, did your +5 take a hit? That could be the
> cause of all your problems.
>
> 3. If it was strapped for 110 and plugged into 220, how much current
> would the transformer itself draw due to the overvoltage? Would that
> pop the fuse all by itself? I've never tried such a thing ( and have no
> intentions of trying it!).
>
> Ed
>
> Brooke Clarke wrote:
>> Hi:
>>
>> I've traced enough of the circuit to know the Zener diodes that blew
>> act as input protection for the A4 power supply. This unit had a
>> sticker saying "Strapped for 220 VAC" when I received it. I changed
>> it to 120 VAC.
>>
>> BUT, if the prior owner plugged it into 220 VAC while it was strapped
>> for 120 VAC this is exactly what you would expect to happen. For
>> partial board layout and partial schematic see:
>> http://www.prc68.com/I/HP3458DVM.shtml#A4
>>
>> Note that the line fuse for 220 VAC is 0.5 Amp and for 120 VAC is 1.5
>> Amp. The Zener diodes are rated for 5 Watts so a drop of 40 volts can
>> only support 125 ma. and the line fuse would be 1.5 Amp so offers no
>> protection. Best to put a fuse at the input to the board as a way to
>> protect A4 from this mistake.
>>
>> Could probably run it without the Zener diodes, but will replace them
>> prior to powering it up.
>>
>> Have Fun,
>>
>> Brooke Clarke
>> http://www.prc68.com
>>
>> Brooke Clarke wrote:
>>> Hi:
>>>
>>> I've started a web page for the HP 3458A and it's problem with more
>>> info at:
>>> http://www.prc68.com/I/HP3458DVM.shtml
>>>
>>> Have Fun,
>>>
>>> Brooke Clarke
>>> http://www.prc68.com
>>>
>>> Brooke Clarke wrote:
>>>> Hi Greg:
>>>>
>>>> I tried both Blue then Error and Blue and Error and could not get
>>>> anything, i.e. the display just showed ISOLATOR DTACK
>>>>
>>>> Have Fun,
>>>>
>>>> Brooke Clarke
>>>> http://www.prc68.com
>>>>
>>>> Greg Burnett wrote:
>>>>> Brooke:
>>>>>
>>>>> ...Correction to my last post... If I remember correctly, you would
>>>>> hit "key shift blue" + "Clear" to clear out each error message in
>>>>> order to read the next error message (if any). You keep clearing
>>>>> the error messages until there are none left. Be sure to use the
>>>>> left & right arrows to scroll through a given error message's
>>>>> complete text line.
>>>>>
>>>>> Poul-Henning's suggestion (about the optical cables on the bottom
>>>>> side) is worth trying. If I remember correctly, the box will still
>>>>> function (except for Ext. Trigger Input and VM Complete Output) if
>>>>> the rear optical cables are disconnected. However you will get an
>>>>> isolator error if the front optical cables are disconnected. But
>>>>> I'm not sure if it will be your error 201 or 200 (Isolator does not
>>>>> respond).
>>>>>
>>>>> Let us know what you find.
>>>>>
>>>>> Greg
>>>>>
>>>>>
>>>>>
>>>>> _______________________________________________
>>>>> time-nuts mailing list -- time-nuts at febo.com
>>>>> To unsubscribe, go to
>>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>>> and follow the instructions there.
>>>>>
>>>>
>>>> _______________________________________________
>>>> time-nuts mailing list -- time-nuts at febo.com
>>>> To unsubscribe, go to
>>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>> and follow the instructions there.
>>>>
>>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
More information about the Time-nuts_lists.febo.com
mailing list