[time-nuts] DMTD question
Magnus Danielson
magnus at rubidium.dyndns.org
Thu Jun 18 00:21:00 UTC 2009
Bruce Griffiths skrev:
> Magnus Danielson wrote:
>> Hi Corby,
>>
>>> Corby
>>>
>>> The noise contribution of the offset oscillator is incompletely canceled
>>> when the delay between the zero crossings for both channels isn't zero.
>>> The effect increases with the delay between zero crossings as your tests
>>> demonstrate.
>>> You can deduce little from the test apart from the fact that Osc 3 has
>>> worse performance than that of the other 2 oscillators.
>> I agree with Bruce, the delay removes the cross-correlation gain in
>> the suppression of the transfer oscillator. Compare the oscillators
>> pair-wise instead.
>>
>> Cheers,
>> Magnus
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com
>> To unsubscribe, go to
>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
> Another question that arises is the stability/jitter of the 100ms delay
> itself.
> How was this 100ms delay achieved?
Indeed. The NBS DMTD system I know of doesn't have this feature.
Cheers,
Magnus