[time-nuts] About HP10544A
Esa Heikkinen
scifiscifi at sci.fi
Sat Mar 7 17:58:14 UTC 2009
Hello!
I got one of the 544's and plan to use it for rubidium cleaning loop
(Efratom LPRO). But I cannot get a clean 10 MHz signal out from it!
There are harmonics and spurious. I know that spurious come from oven
swicther so those should be easily handled with separate oven supply.
But the bigger problem are the harmonics:
http://www.amigazone.fi/files/gpsdo/544-4.png
This spectrum is measured with emitter follower buffer connected to
output which is terminated with 1k like datasheet recommends. So the
load should be ideal but still distorted. The distortion can be seen
even with oscilloscope:
http://www.amigazone.fi/files/gpsdo/544-6.png
So it ooks like negative half cycle is distorted.
As a reference, THIS is the signal to be cleaned with HP10544A:
http://www.amigazone.fi/files/gpsdo/LPRO-3.PNG
:-)
So... I believe that there may be something wrong with HP10544A if it's
output is worse than lower grade xtal oscillators used in LPRO etc.
But just to make sure I'm asking you guys having these oscillators.
Should it even give clean sine wave or not? And if it should, where
should I start to seek the fault?
--
73s!
Esa
OH4KJU
More information about the Time-nuts_lists.febo.com