[time-nuts] Reverse isolation

Bruce Griffiths bruce.griffiths at xtra.co.nz
Sun Mar 8 23:18:07 UTC 2009


Didier

NIST have measured reverse isolation as high as -144dB @ 10MHz for one
of their isolation amplifier designs but they gave no measurement system
details.

Bruce

Didier Juges wrote:
> As long as one is not trying to measure extremely low signals (or extremely
> high isolation), the conventional method works, and I have used it many
> times, I am not sure that the method would scale when dealing with very high
> levels of isolation.
>
> I certainly would not trust a conventional VNA for isolation of 100dB or
> more, simply because such isolation is difficult to achieve in an instrument
> that is supposed to be able to switch its source and receivers between the
> two ports.
>
> The HP 8722D which we use at work is only specified with 100dB dynamic
> range, and I am not sure how that applies to reverse isolation.
>
> Using a slightly offset test frequency while driving the device at its
> normal operating frequency would only work when using a spectrum analyzer as
> receiver. Most VNAs don't have the capability of rejecting large signals
> close in, and separating the frequencies too much would make the test
> invalid with a narrow band amplifier. I have used that method to measure the
> hot output VSWR of a TWT amplifier for instance (that was interesting). Of
> course, the spectrum analyzer does not give you the phase, but it's better
> than nothing.
>
> Didier KO4BB
>
>   
>> -----Original Message-----
>> From: time-nuts-bounces at febo.com 
>> [mailto:time-nuts-bounces at febo.com] On Behalf Of John Day
>> Sent: Sunday, March 08, 2009 5:47 PM
>> To: Discussion of precise time and frequency measurement
>> Subject: Re: [time-nuts] Reverse isolation
>>
>> At 06:14 PM 3/8/2009, Didier wrote:
>>     
>>> This question is directed at Bruce, but if anyone else has a 
>>> contribution, feel free to speak.
>>>
>>> Bruce,
>>>
>>> What is the best way to measure the reverse isolation of an 
>>>       
>> amplifier 
>>     
>>> (particularly a buffer amplifier for a 10 MHz reference), when it is 
>>> expected to be in the order of 100 dB or more?
>>>
>>> Feeding the output with a known signal and measuring at the 
>>>       
>> input with 
>>     
>>> a spectrum analyzer comes to mind, but I am sure there must be 
>>> something wrong with that technique, it sounds too simple.
>>>       
>> No, not really. How else would you measure it? Reverse 
>> isolation is basically the reverse gain - S12 - of the 
>> amplifier. How does a VNA measure S12? Essentially inject a  
>> signal at port 2 and see how much comes out of port 1.
>>
>> If you want to get so picky as to determine S12 with a signal 
>> in the forward direction then you have a problem. Because the 
>> forward gain -
>> S21 - is going to effectively swamp the signal going the other way. 
>> So this is almost impossible to measure if the signals are at 
>> the same frequency. In this case measure the S-parameters of 
>> the amplifier in its 'normal ' configuration, then de-embed 
>> the S-parameters of the device. Assuming the device is the 
>> only non-linear element you are dealing with then from the 
>> S12 & S22 values you can also figure out how the device 
>> reacts in the reverse direction.
>>
>> Then if you don't mind solving a large matrix you can figure 
>> out how the device might react to passing a signal in both 
>> directions. The reality is however that if the device is 
>> within its linear range, which it is likely to be if you want 
>> to distribute a reference or some such, the reverse behaviour 
>> of the well terminated amplifier will approach the nominal 
>> S12 value. The difficult part is to determine what happens 
>> when the amplifier is not nicely terminated!
>>
>> John
>>
>>
>>     
>>> The presence of a signal at the input (or not) may affect 
>>>       
>> the operating 
>>     
>>> point of the amplifier, so measuring from output to input 
>>>       
>> without such 
>>     
>>> signal may not give a true result.
>>>
>>> Didier
>>>
>>>
>>>
>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts at febo.com To unsubscribe, go to 
>>> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>>
>>>
>>> --
>>> No virus found in this incoming message.
>>> Checked by AVG.
>>> Version: 7.5.557 / Virus Database: 270.11.9/1989 - Release Date: 
>>> 3/7/2009 6:43 PM
>>>       
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com To unsubscribe, 
>> go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
>>
>> --
>> No virus found in this incoming message.
>> Checked by AVG. 
>> Version: 7.5.557 / Virus Database: 270.11.9/1989 - Release 
>> Date: 3/7/2009 6:43 PM
>>
>>     
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>   





More information about the Time-nuts_lists.febo.com mailing list