[time-nuts] 5370A
Joseph Gray
jgray at zianet.com
Sat Oct 10 08:17:38 UTC 2009
Oh, one last question for the night. I noticed that the CPU board has
an emply ROM socket. I assume this is normal?
OK, let's make it two questions. While I'm working on this thing,
should I pull the EPROMS and ROMS and read them? I've heard of EPROMS
going bad in some other old equipment (maybe it was on this list). Of
course, finding replacement chips might be a challenge.
Joe Gray
KA5ZEC
On Sat, Oct 10, 2009 at 2:11 AM, Joseph Gray <jgray at zianet.com> wrote:
> Bruce,
>
> Thanks for the tip. I was very careful. Only one board was really
> stuck. I left it alone for now.
>
> Just for the heck of it, I did a few more steps in the Chapter 4
> Operation Verification. As I said, discounting the bad BNC, step 13
> reads aproximately 103.6nS, so I continued.
>
> Steps 14, 15 and 16 are ok.
>
> Step 17 shows 3.6nS (should be < 1nS). Period Compliment shows
> -96.3nS. Something very wrong here.
>
> Steps 18, 19 and 20 are ok.
>
> I'm going to stop there for the night and drag out the scope tomorrow
> to pickup at step 21.
>
> Thanks for the help so far.
>
> Joe Gray
> KA5ZEC
>
> On Sat, Oct 10, 2009 at 1:50 AM, Bruce Griffiths
> <bruce.griffiths at xtra.co.nz> wrote:
>> Joe
>>
>> Just be careful no to break the plastic handles off the boards.
>> The board retention force can be quite high.
>>
>> Bruce
>>
>> Joseph Gray wrote:
>>> Bruce,
>>>
>>> I have a 100MHz scope, DVM, frequency counter, signal generator, etc.
>>> Sorry, no spec an. I wish I had one.
>>>
>>> The first problem is that it fails the first performance test. Instead
>>> of 100nS, I get about 10nS and that slowly increases to about 12nS.
>>> This is with the unit already warmed up and it is repeatable. I
>>> checked the internal reference output and it is very close to 10MHz,
>>> so that isn't the problem.
>>>
>>> Before I go any further, I think I'll open it up reseat all the
>>> boards. It couldn't hurt.
>>>
>>> Joe
>>>
>>> On Fri, Oct 9, 2009 at 11:51 PM, Bruce Griffiths
>>> <bruce.griffiths at xtra.co.nz> wrote:
>>>
>>>> Joseph Gray wrote:
>>>>
>>>>> I just got a 5370A locally and am going through the performance
>>>>> checks. Things are not as they should be. Is there someone who knows
>>>>> this unit intimately who would be willing to help me off list?
>>>>>
>>>>> Joe Gray
>>>>> KA5ZEC
>>>>>
>>>>>
>>>>>
>>>> Joe
>>>>
>>>> What do you have in the way of test gear?
>>>> A 100+MHz scope would be useful as well as a spectrum analyser.
>>>>
>>>> Bruce
>>>>
>>>>
>>>> _______________________________________________
>>>> time-nuts mailing list -- time-nuts at febo.com
>>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>>> and follow the instructions there.
>>>>
>>>>
>>>>
>>>
>>> _______________________________________________
>>> time-nuts mailing list -- time-nuts at febo.com
>>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>>> and follow the instructions there.
>>>
>>>
>>
>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>>
>>
>
More information about the Time-nuts_lists.febo.com
mailing list