[time-nuts] 10811 changing frequency when load impedance changes ?
Mark Spencer
mspencer12345 at yahoo.ca
Sat Dec 25 06:42:35 UTC 2010
Thanks..
I'll likely do some more testing with the 50 ohm load in place and see how
stable this unit is before investing more time in this.
----- Original Message ----
From: Bruce Griffiths <bruce.griffiths at xtra.co.nz>
To: Discussion of precise time and frequency measurement <time-nuts at febo.com>
Sent: Fri, December 24, 2010 10:05:51 PM
Subject: Re: [time-nuts] 10811 changing frequency when load impedance changes ?
Mark Spencer wrote:
> Seasons greetings.
> I was pleasantly surprised to find that a new to me HP 5328A had an 10811 in
> it. (I was expecting the older OCXO..) After leaving it running for a
> week I adjusted the frequency and while doing so I noticed that the out put
> frequency changes by a third of a hz or so if I add or remove a 50 ohm load.
>I'm measuring the frequency at the timebase output connection of the
> 5328A. Is this to be expected ?
>
> Many thanks
> Mark Spencer
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
>
No, that seems to be a couple of orders of magnitude larger than the specs for
the 10811A would lead one to believe.
Maybe there's a problem with the output buffer amp in the 10811A?