[time-nuts] ADEV vs MDEV

Bruce Griffiths bruce.griffiths at xtra.co.nz
Sat Feb 6 20:11:13 UTC 2010


Sounds good but you still haven't found its Achilles heel:

The frequency measures need to be integrated (either implicitly or 
explicitly) to produce phase measures which can then be used to 
calculate ADEV, MDEV etc.
The major problem is that integration amplifies the small errors that 
are inevitably present.
In practice (except for very noisy sources) the technique isnt 
particularly useful for Tau more than a few times the inverse PLL bandwidth.

Thus NIST and others quietly dropped this method several decades ago.
This is alluded to in Steins recent paper availble on the Symmetricom 
website:

*The Allan Variance – Challenges and Opportunities*


Bruce

WarrenS wrote:
> Peat said:
>    
>> I would appreciate any comments or observations on the topic of apparatus with demonstrated stability measurements.
>> My motivation is to discover the SIMPLEST scheme for making stability measurements at the 1E-13 in 1s  performance level.
>>      
>
> If you accept that the measurement is going to limited by the Reference Osc,
> for Low COST and SIMPLE, with the ability to measure ADEVs at that level,
> Can't beat a simple analog version of  NIST's "Tight Phase-Lock Loop Method of measuring Freq stability".
> http://tf.nist.gov/phase/Properties/one.htm#oneone    Fig 1.7
>
>
> By replacing the "Voltage to freq converter, Freq counter&  Printer with a Radio shack type PC data logging DVM,
> It can be up and running from scratch in under an Hr, with no high end test equipment needed.
> If you want performance that exceeds the best of most DMTD at low Tau it takes a little more work
> and a higher speed oversampling ADC data logger and a good offset voltage.
>
> I must add this is not a popular solution (Or a general Purpose one) but
> IF  you know analog and have a GOOD osc with EFC to use for the reference,
> as far as I've been able to determine it is the BEST SIMPLE answer that allows High performance.
> Limited by My HP10811 Ref OSC, I'm getting better than 1e-12 in 0.1 sec (at 30 Hz Bandwidth)
>
> Basic modified NIST Block Diag attached:
> The NIST paper sums it up quite nicely:
> 'It is not difficult to achieve a sensitivity of a part in e14 per Hz resolution
> so one has excellent precision capabilities with this system.'
>
> This does not address your other question of ADEV vs MDEV,
> What I've described is just a simple way to get the Low cost, GOOD Raw data.
> What you then do with that Data is a different subject.
>
> You can run the raw data thru one of the many ADEV programs out there, 'Plotter' being my choice.
>
>
> Have fun
> ws
>
> *************
>
> [time-nuts] ADEV vs MDEV
> Pete Rawson peterawson at earthlink.net
> Sat Feb 6 03:59:18 UTC 2010
>
> Efforts are underway to develop a low cost DMTD apparatus with
> demonstrated stability measurements of 1E-13 in 1s. It seems that
> existing TI counters can reach this goal in 10s. (using MDEV estimate
> or 100+s. using ADEV estimate). The question is; does the MDEV tool
> provide an appropriate measure of stability in this time range, or is
> the ADEV estimate a more correct answer?
>
> The TI performance I'm referring to is the 20-25 ps, single shot TI,
> typical for theHP5370A/B, the SR620 or the CNT81/91. I have data
> from my CNT81showing MDEV<  1E-13 in 10s. and I believe the
> other counters behave similarly.
>
> I would appreciate any comments or observations on this topic.
> My motivation is to discover the simplest scheme for making
> stability measurements at this performance level; this is NOT
> even close to the state-of-the-art, but can still be useful.
>
> Pete Rawson
>
>    
>
>