[time-nuts] HP 5371/2 error

Mark Sims holrum at hotmail.com
Sun Mar 21 01:37:41 UTC 2010


Basically it means the unit has lost the battery backup to the calibration memory.   

The cure is a rather simple calibration step.  On the 5372A you input a 60 mV 10KHz sinewave and perform Test 25.

The adjustment is also called "Maximum Hysteresis" 		 	   		  
_________________________________________________________________
Hotmail: Trusted email with powerful SPAM protection.
http://clk.atdmt.com/GBL/go/210850553/direct/01/


More information about the Time-nuts_lists.febo.com mailing list