The 100 ohm standard for RTD's dates way back. The assumption was that you had it on a *long* run of cable (2 pair / sense leads of course). The insulation leakage was a bigger issue than anything else in the equation.
Bob
On Nov 12, 2010, at 5:36 AM, Poul-Henning Kamp wrote:
> In message <20101112110627.488cbc8f at vz127.worldserver.net>, "Florian E. Teply"
> writes:
>
>> In a bridge circuit, you don't measure resistance directly, but use the
>> voltage that appears across the bridge. So for a 100 ohms element,
>> you'd usually have ten times the current flowing in that branch
>> compared to a 1kohm element.
>
> This was one of the things that I wondered about: How large currents
> are used ?
>
> Can't be too much because that would lead to self-heating...
>
> Poul-Henning
>
> --
> Poul-Henning Kamp | UNIX since Zilog Zeus 3.20
> phk at FreeBSD.ORG | TCP/IP since RFC 956
> FreeBSD committer | BSD since 4.3-tahoe
> Never attribute to malice what can adequately be explained by incompetence.
>