[time-nuts] Questions about HP 5370B
Greg Burnett
gbusg at comcast.net
Tue Sep 14 08:11:15 UTC 2010
Hi Charles,
Yes, the rear heatsink does run very hot, approx. 61 deg C. But do make sure
the fan is running! The unit will be damaged without the fan. Also it's
important to refrain from running the instrument too long with its top cover
removed, because the fan air doesn't properly loop around and blow back out
the heatsink if you've left the instrument top cover off.
As Bruce suggests, you'll want to peek inside to see that you really do have
a 10811 oscillator. If so, then it sounds like the thermal fuse (F1, inside
the 10811) might be open. According to the manual, I think the HP part
number for F1 was 2110-0617. However, Agilent's current online parts search
does not find that p/n. Maybe you'll have to call Agilent's "Parts ID" to
identify the current part number for F1? ...Or probably others in this
forum will offer more current information regarding what part to use for F1.
Greg
More information about the Time-nuts_lists.febo.com
mailing list