[time-nuts] HG414A/CH1-78 report
Magnus Danielson
magnus at rubidium.dyndns.org
Sun Apr 3 03:29:35 UTC 2011
On 04/02/2011 06:53 PM, Magnus Danielson wrote:
> Hi!
>
> Yesterday I managed to trim away the unwanted RF from the 5 MHz output.
> I fail to get as good result from the 10 MHz output, but I think it will
> not be as big impact. Will possibly need to handle that. Spectrum
> analyzer helps a lot more than a scope. Happy I invested in a faster
> analog scope.
>
> It is not a stable lock, even if I can se it lock in. I see too much
> noise in it and the status flickers so I think I can blame the lamp
> detection, need to check that. Possibly I will have to do the lamp
> rejuvination trick. Also I need to figure out how to do the disassembly,
> since I am not yet clear on how to best reach every part.
OK. Once I figured out how the side-plate slided off I was able to
unscrew the rubidium physical package from the Aluminium frame.
It opens up somewhat like a russian doll and it is clear that it is a
combination of magnetic shielding and thermal isolation in layers. In
fact, it seems not to draw a lot of current. Once I got sufficiently far
in I was able to pull out the rubidium lamp assembly. It's a huge thing,
by far the largest rubidium set up I've seen so far with the Rubidium
lamp being a double cavity thing creating a 1 cm long and maybe 4 mm
wide rubidium discharge then being reflected by a copper reflector to
form a 35 mm wide optical path through the other two glass assemblies.
The lamp lits up and it is not that bad, but I need to think. There are
many possible causes for a bad 2nd harmonic response, and rubidium lamp
may be just one possible problem.
Time for bed.
Oh, I *did* try to get the camera. I discovered it badly needed the
battery to recharge, so it is charging it now.
Cheers,
Magnus