[time-nuts] T.I. experimenting - newbie question
Hal Murray
hmurray at megapathdsl.net
Tue Apr 5 04:21:27 UTC 2011
> Here are the results I got using your suggestion. The numbers may be a bit
> different than last night as I'm not sure I'm using the same cable.
> A+, B+ = 18.9 ns
> A-, B- = 19.4 ns
> Obviously there is some difference in delay between the A and B channels.
> Otherwise, the two numbers would have been identical - correct?
...
> If I take the average of the two readings, I get 19.05 ns, which is more
> precise than the readings I'm taking. I'm rounding to the nearest 0.1 ns on
> the readings.
There are several possibilities for asymmetry.
The input signal could have other than a 50/50 high/low ratio.
The trigger levels could be set at other than the half-way point.
Do you have a scope?
There are probably second order problems with different rise/fall times.
> A+, B- = 5014.6 ns
> A-, B+ = 5023.5 ns
> If I take the average of these two readings and subtract out the 5 us for
> 1/2 period of the 100 KHz square wave, if get 19.15 ns. This agrees
> very closely with the above average.
I read that as everything mostly works.
The question is how well does it work and/or what can you do to get
more/better data?
You can get another set of numbers by swapping the A and B inputs so the
signal goes from B to A rather than A to B.
Some of the errors might be more obvious if you deliberately offset the
trigger levels.
NB: That's for all 4 modes:
A+ to B+ will be slightly less than a whole cycle.
A+ to B- will be slightly less than a 1/2 cycle.
...
--
These are my opinions, not necessarily my employer's. I hate spam.