[time-nuts] Question for any HP105 owners HP5370 Settings
Paul A. Cianciolo
paulc at snet.net
Wed Aug 31 00:37:12 UTC 2011
John,
I don't know if I am going to understand all this.
I come from an analog world, a dinosaur, all this timing stuff as much fun
as it is, can be awful intimidating.
I am thinking there is a 1 PPS source in the Z3801a but for the moment I
have one of these devices
http://cgi.ebay.com/ws/eBayISAPI.dll?ViewItem&item=320751663786
Do you think this will work?
Some day when you have some free time, perhaps we could have a phone call or
SKYPE if you use that and discuss 1 or 2 of these issues
Thank you again
Paul A. Cianciolo
W1VLF
http://www.rescueelectronics.com/
Our business computer network is powered exclusively by solar and wind
power.
Converting Photons to Electrons for over 20 years
-----Original Message-----
From: time-nuts-bounces at febo.com [mailto:time-nuts-bounces at febo.com] On
Behalf Of John Miles
Sent: Tuesday, August 30, 2011 3:01 PM
To: 'Discussion of precise time and frequency measurement'
Subject: Re: [time-nuts] Question for any HP105 owners HP5370 Settings
> 1) Under Function
> The counter is in the "TI" mode
>
> 2) Under statistics mode
> The counter is in the "Mean" mode
>
> 3) Under Sample Size
> Number "1" is selected
>
> 4) Under arming
> The "+TI" button is selected.
>
> 5) Under display rate.
> Fully clockwise the display nor the GPIB interface will update if
> this is in the "hold" position
>
> 6) Inputs section
> All three pots are in the "preset" position.
> 10 MHz Z3801A Reference is connected to "Start" DUT is connected to
> "Stop
> Start Stop inputs are "positive slope" and each is "50 Ohms"
> Termination Both input switches are in the "divide by 1 " position.
> External BNC has no input.
>
> External reference drive, in the back = the Z3801A
It's probably better not to use one of your test sources as the counter's
external reference. That's asking for systematic-bias trouble. The
internal 10811 reference is quieter than the rest of the counter for
measurements less than ~10 seconds apart, so there is rarely a good reason
to use an external reference at all, unless you are looking to measure
absolute frequency.
> LED displays " 53.20 with the. 20 changing N secs"
> Remote led lit
> "Talk" in the GPIB area is lit.
What limits the rate at which readings are coming out of the counter? You
need to either back off the rate control to yield about 10 readings/second
or fewer (and let TimeLab measure the interval), or use a 1-pps or 10-pps
source on the START or external-trigger jacks. Otherwise I'm surprised the
GPIB-Ethernet adapter is keeping up... and at best you are likely to get
irregular tau intervals.
If/when you go to a pulsed source on the START jack, don't leave the level
control in its preset position, or you'll probably get some extra noise near
0V. Use DC triggering and set the level about halfway up the edge (e.g.,
with 5V pulses in from a CMOS divider, you could set ~2V trigger level in
x1 or 0.25V in x2, whichever gives you the lowest residual noise.)
-- john, KE5FX
_______________________________________________
time-nuts mailing list -- time-nuts at febo.com To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list