[time-nuts] HP 5372A vs. 5370A
Ed Palmer
ed_palmer at sasktel.net
Wed Feb 9 21:14:03 UTC 2011
If you've got one, it can be hard to justify buying the other unless you
have some very specific requirements. There's a lot of overlap between
them. I have a 5372A and have wondered if it's worth getting a 5370(A
or B). So far, the answer for me is no.
Here are a few things that I've used the 5372A for.
I have a Navsync CW-12 GPS receiver. I measured 1000 periods of the
1PPS and saw that the Max-Min range is less than 30ns and the Standard
Deviation is less than 5 ns. The minimum, maximum, Mean, RMS, and
standard deviation are automatically calculated and shown by the 5372A.
By the way, my TBolt has a range of < 4 ns and a Std Dev of < 1 ns.
The CW-12 also has a 10 MHz output. I used the Histogram Time Interval
function to measure the periods of 100 Million cycles with a resolution
of 200ps. It took less than 30 seconds to measure, process, and display
the results. The results showed that there was a normal distribution
around 100 ns and a second normal distribution around 92 ns. The
difference is approximately the period of the internal clock (120 MHz).
This told me that the 10 MHz is kept on frequency by occasionally
shortening the period by one cycle of the internal clock. For my unit,
this happens about 200 times a second. I know this by the ratio of the
number of long periods to short periods. This behaviour explains why
Navsync warns that this output needs to be cleaned up before using it as
a frequency reference.
I routinely use the 5372A to make multiple 1 sec. measurements of
frequency or time interval and then dump the results via GPIB for
analysis. Standard stuff for either the 5370 or 5372 except as
mentioned elsewhere, these measurements are made with no dead time.
My unit has Option 040 which adds FFT capabilities. I can do phase
noise measurements within the limitations of the 200ps resolution of the
box.
The June 2008 issue of Circuit Cellar had an article ("Let's be Crystal
Clear") on crystals and oscillators. The 5372A was used to analyze the
startup behaviour of an oscillator. You should be able to get this
article through inter-library loan or maybe there's a copy lurking on
the net somewhere. There are also various manuals and app notes online
at the Agilent site.
The biggest limitation of the 5372A is that you can only make 8191
frequency measurements or 4095 time interval measurements with each
measurement lasting no longer than 8 seconds. If you want to use it for
making more or longer measurements (e.g. long term Allan Deviation
measurements) you have to start doing some GPIB programming, but I think
you lose the no-dead-time advantage.
If you see a 5372A that's reporting an "Error 160: Out of sensitivity
cal." don't worry about it. It just means that the onboard lithium
battery is dead and needs to be replaced. The calibration is trivial to do.
The most important thing about either the 5370 or 5372 is to study it.
Figure out it's capabilities, limitations, and tricks so that you use it
to it's full potential. I'm still working on that!
Ed
Bert, VE2ZAZ wrote:
> Hi everyone.
>
> I own an HP 5370A. I do GPIB and Labview, KE5FX Utilities, DF6JB's Plotter,
> spreadsheets etc. Always looking to improve my setup...
>
> Here is a simple question: Why should I try to get an HP 5372A (or 5371A)? What
> are the benefits over the 5370A? Worth spending?
>
> Thanks in advance,
>
> Bert.
>
>
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
>
More information about the Time-nuts_lists.febo.com
mailing list