The truth probably lies somewhere in between.
Joe
-----Original Message-----
From: time-nuts-bounces at febo.com [mailto:time-nuts-bounces at febo.com] On
Behalf Of beale
Sent: Thursday, March 17, 2011 1:42 PM
To: time-nuts at febo.com
Subject: [time-nuts] semi-OT: comparison of two 10x scope probes (cost $7
vs. $406)
I am always loosing the little hook adaptor on the end of my scope probes
so I went looking for replacements. I noticed that I could buy a pair of
complete "100 MHz" probes including tip and ground clip online for $13
shipped from Shenzen China. Didn't expect much, but this application was for
low speed signals anyway and I got some. Couldn't loose at that price. I
figured the hook tip adaptor alone was worth it.
They arrived the next week and just for fun I compared my new $6.50 probe
with a 500 MHz Tek P6139A probe, using a not terribly fast scope, the 100
MHz Tek TDS 220, and a signal of unknown risetime, the output of a 26 MHz
Pletronics OCXO in a 14DIP can (by the way these go for $2 online, the
cheapest OCXO I've seen). This was just a quick casual comparison, not to be
taken too seriously.
I was somewhat surprised to see the cheap probe presented the same 3.8 ns
risetime as the Tek probe on this scope, although it comes with some ringing
also whereas the Tek probe did not show ringing. Both probes were used in
the same place, ground directly to probe shield without flying ground leads
(although the circuit itself was casually constructed and had about 1"
length of ground return). If you had a simple 1-pole 100 MHz lowpass you'd
expect 3.5 nsec risetime so this is in the right ballpark for scope rated at
100 MHz (and my source signal may be no faster than that anyway- I haven't
built an avalanche pulse generator yet :-).
Scope photo showing the comparison below. The cheap probe trace is shown as
RefA (light grey) and the Tek probe is CH1 (dark grey/black).
https://picasaweb.google.com/lh/photo/1_SVMQ3ajYJUARc51BXnzw?feat=directlink
_______________________________________________
time-nuts mailing list -- time-nuts at febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
-----
No virus found in this message.
Checked by AVG - www.avg.com
Version: 10.0.1204 / Virus Database: 1498/3513 - Release Date: 03/17/11
More information about the Time-nuts_lists.febo.com