[time-nuts] Question about an HP5370B TIC
John Miles
jmiles at pop.net
Fri Sep 30 07:56:01 UTC 2011
> > If I put the 5370B in the +/-TI mode I expected to see both + and -
small
> > time differences when the two signals are closely aligned - and mainly
I
> > do. Typically the difference is + or - a usec or less. Occasionally
> > however the 5370B displays very near to a +1 or -1 second (999.999 xxx
ms
> > or -1.000 000 xxx s for example) reading. Needless to say this plays
havoc
> > with trying to measure a mean value of the alignment. The 5370B is
clearly
> > measuring the "outside" between the pulse pair when I want it measure
the
> > "inside" (or short time interval) between the pulse pair.
>
> How are you collecting the data? If each sample gets to a PC, you can fix
it
> up in software.
> If X > 0.5: X = X -1.0
> If X < -0.5: X = X+1.0
> One disadvantage of that is that those samples eat 2 pairs of pulses so
you
> are only getting half as as many samples per hour.
>
> I like John's suggestion of triggering one side off the other edge of the
> pulse.
>
> Is the pulse from the OCXO supposed to be synchronized with the GPS
second?
> If not, just zap the counter to move the pulse around within the one
second
> window. (I'm thinking of shorting out one of the counter bits with a
> screwdriver or clip lead for a fraction of a second.)
>
> My first thought was to insert a delay line, but that will take 1000 feet
of
> coax. Maybe more. I don't have that handy. Frys says $80 or $90 for 500
> feet. I'm too cheap to buy it just for one test like this. (But I might
> use this as an excuse if I see some at a good price.)
The easy way to 'cheat' may be to lie to your GPS clock's control software
and tell it you have several hundred feet of antenna cable. Most commercial
clocks will try to maintain accurate time vis-a-vis the antenna site, so
they can delay the pulse to compensate.
Or you can just let it wrap, and compensate in software...
-- john, KE5FX