[time-nuts] OT: Question on dead HP34401
Ulrich Bangert
df6jb at ulrich-bangert.de
Mon Dec 17 07:57:28 UTC 2012
Gentlemem,
pardon for the delayed answer, I had some extreme busy days the last week.
> The floating logic is reset by the LM2925 5V regulator reset
> output. There is no connection to this reset line aprt from
> the Gate array reset
> input and the micro (80196) reset input.
Yes, U553 on page 9-15 has this reset output which is called XPONRST. After
it has passed the lowpass R503/C504 on page 9-13 (upper left) it enters U500
and U501. I observe this 7 kHz reset signal RIGHT from the lowpass even if I
lift up the voltage regulator's reset output pin.
Best regards
Ulrich Bangert
> -----Ursprungliche Nachricht-----
> Von: time-nuts-bounces at febo.com
> [mailto:time-nuts-bounces at febo.com] Im Auftrag von Bruce Griffiths
> Gesendet: Dienstag, 11. Dezember 2012 22:46
> An: Discussion of precise time and frequency measurement
> Betreff: Re: [time-nuts] OT: Question on dead HP34401
>
>
> It only resets the earth referenced micro (8051) on power up and when
> pulses on the optocoupler output arent present.
> There is no other connection to the reset pin of this micro.
>
> The floating logic is reset by the LM2925 5V regulator reset
> output. There is no connection to this reset line aprt from
> the Gate array reset
> input and the micro (80196) reset input.
>
> The relay drivers are reset by a 80196 output pin.
>
> The display logic is reset by another 90196 output pin.
>
> Bruce
>
> paul swed wrote:
> > OK but why would it tie to the reset line?
> >
> > On Tue, Dec 11, 2012 at 3:48 PM, Marco IK1ODO
> -2<ik1odo at spin-it.com>
> > wrote:
> >
> >
> >> At 21:40 11/12/2012, you wrote:
> >>
> >>
> >>> Two comments the manuals available on agilent
> >>> What on earth is a crossguard?
> >>>
> >>>
> >> Data exchange between a floating, guarded circuit and a ground
> >> referenced one?
> >>
> >> Marco IK1ODO
> >>
> >>
> >>
> >>
> >> ______________________________**_________________
> >> time-nuts mailing list -- time-nuts at febo.com
> >> To unsubscribe, go to https://www.febo.com/cgi-bin/**
> >>
> mailman/listinfo/time-nuts<https://www.febo.com/cgi-bin/mailman/listi
> >> nfo/time-nuts>
> >> and follow the instructions there.
> >>
> >>
> > _______________________________________________
> > time-nuts mailing list -- time-nuts at febo.com
> > To unsubscribe, go to
> > https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> > and follow the instructions there.
> >
> >
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.