The ADEV plot was done after the unit had fully settled to it's new location in front of the 53132A fan.
The Tsc5125A instrument has a limit of 9 minutes to plot the frequency drift for some reason, so I can only show you the last 9 minutes.
The transient due to the new temperature was long gone when I started this ~20 min ADEV plot.
Bye,
Said
Sent from my iPad
On Dec 20, 2012, at 4:36, Bob Camp <lists at rtty.us> wrote:
> Hi
>
> Did I miss part of the data?
>
> The plot I got shows about 9 minutes. Temperature step stress takes a *long* time to settle out.
>
> Bob
>
> On Dec 19, 2012, at 9:56 PM, SAIDJACK at aol.com wrote:
>
>> Not sure about that, if you look at the frequency plot after ~20 minutes in
>> moving air the frequency was still extremely close to 10.000000MHz.. to
>> within <1E-011 of 10MHz. This is a free-running 10811.
>>
>> Compare that to the plot of the OCXO I had sent out some hours ago when it
>> was running stable inside the foil - there was almost no average frequency
>> change between the two tests.
>>
>> If the heaters had an issue keeping average crystal temperature stable,
>> then the frequency would have changed from the first plot to the last plot I
>> would think. In my opinion the airflow is just adding a huge bunch of heater
>> control loop noise to the output stability, or there are components on the
>> PCB which are very sensitive to the airlfow. Consider that this unit and
>> it's PCB was designed to live inside the 53132A (very close to the fan) that
>> I am now using as an air source. One thing this tells me: the fan in the
>> counter could be disabled and it would improve the units stability, if one
>> keeps an eye on internal counter temperature.
>>
>> bye,
>> Said
>>
>>
>> In a message dated 12/19/2012 18:42:08 Pacific Standard Time,
>> actast at hotmail.com writes:
>>
>> I think the data shows that the heaters were losing ground, which would
>> explain the steadily falling temp of the SC cut quartz.
>>
>> Thomas Knox
>>
>>
>>
>> _______________________________________________
>> time-nuts mailing list -- time-nuts at febo.com
>> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
>> and follow the instructions there.
>
>
> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list