[time-nuts] ensemble oscillators for better stability
cdelect at juno.com
cdelect at juno.com
Fri Dec 28 19:12:26 UTC 2012
Hi,
I've been playing around with 2 HP 5065A standards to see if averaging
them will give better stability.
The results so far are:
1 sec 10 sec 100 sec
unitA 1.28-12 3.83-13 1.33-13
unitB 9.31-13 3.44-13 1.18-13
combined 8.95-13 3.09-13 8.95-14
Each unit was adjusted for equal amplitude into a resistive power
combiner and also adjusted so their phase was coincident prior to
starting the plots.
The main problem was getting their relative frequencies close enough so
that there was no drift during the measurements.
Not a trivial accomplishment for the 100 sec tau!
I'm thinking now of a way to loosely lock the phase of one unit to the
other to simplify the setup.
One different idea I had that I'd like some feedback on:
If I take both units into a double balanced mixer, then take the
resulting 10MHz and divide it back to 5MHz would I get the same
improvement?
This method would be much simpler!
Happy New Year!
Corby Dawson
More information about the Time-nuts_lists.febo.com
mailing list