[time-nuts] Efratom FRK - no 10Mhz oscillator
Don Lewis
dlewis6767 at austin.rr.com
Wed Jan 4 17:35:24 UTC 2012
Thanks, all, for your recent help on my Efratom. I finally got it
operational.
The crystal-holder in the [hot] OCXO was bad; one pin looked like it had
corroded in two.
I had to dig out the crystal and holder (it was REALLY potted with grey
RTV.)
Once I got it all back together, I had to tweak a 'factory-set' small pf
capacitor in the VCO/varactor circuit; and then it locked after about 5
minutes.
Since it is now my only 'standard' I guess I will proclaim myself a true
TIME-NUT and start seeking an even better standard. I have three Rockwell
GPS-D200 modules ...maybe tying all three together someway will help me vet
the rubidium.
Btw ...Is there any place in Austin, TX I could walk in with the Efratom and
get a 'cesium' check on the Efratom and calibrate it even further???
Appears it has a trimmer for such.
-Don
Austin, TX
...........................................
-----Original Message-----
From: time-nuts-bounces at febo.com [mailto:time-nuts-bounces at febo.com] On
Behalf Of Ed Palmer
Sent: Wednesday, December 21, 2011 12:12 PM
To: Discussion of precise time and frequency measurement
Subject: Re: [time-nuts] Efratom FRK - no 10Mhz oscillator
Hi Don,
I'm assuming that you've checked for obvious things like bad solder
joints. After 20 or more years in that heat, anything is possible.
Luckily, the oscillator board doesn't have too many connections to
remove when you pull the board. I use a 20X magnifier to really get
close. I've often found problems that you can't see with less
magnification.
In one message you mentioned that touching the crystal oven with an
external 10 MHz signal starts the oscillator. Isn't the oven supposed
to be grounded?
I agree with other posters that if you have a bad crystal, you'll have
to find a replacement that's made to be used in an oven. However, any
old 10 MHz crystal will be useful to diagnose the problem. It might be
possible to sacrifice another OCXO and transplant the crystal into the
FRK. Your best bet would be an OCXO with an AT crystal. I wouldn't try
using an SC crystal. The circuitry would be significantly different.
The oven would have to be retuned to match the new crystal, but that's
just a resistor change.
If you're up for a bigger challenge, you could try interfacing an
external OCXO to the Rubidium loop. Rubidiums are typically marketed
for low drift applications rather than low Allan Deviation or phase
noise. I've wondered whether a high quality SC-crystal OCXO would make
a significant improvement in a Rubidium's short-term performance. I've
never had a Rubidium with a dead oscillator to experiment on.
Ed
On 12/21/2011 8:04 AM, Don Lewis wrote:
> Thanks, Pete.
>
> Good tips.
>
> I ordered 10 crystals this morning, ...they are dirt cheap. Will try and
> test one as you suggested.
>
> If it is the 'crystal', a bad crystal, .......what then?
>
> Do I try to place one of these cheap ones in the oven, ...or do I try to
> just run it outside the oven or find a high-temperature crystal???
>
> The crystal oven/controller seems fine, ...hot enough to hurt my fingers.
>
> Thanks, ...-Don
>
>
> ------------------------------------------
>
>
>
> -----Original Message-----
> From: time-nuts-bounces at febo.com [mailto:time-nuts-bounces at febo.com] On
> Behalf Of Peter Bell
> Sent: Wednesday, December 21, 2011 5:28 AM
> To: Discussion of precise time and frequency measurement
> Subject: Re: [time-nuts] Efratom FRK - no 10Mhz oscillator
>
> Yeah, that was why I suggested just sticking a random 10MHz xtal
> across the pins on the board (with the existing xtal wiring
> disconnected) - if the rest of the circuit is good, then the
> oscillator should start up with no problems - if it doesn't, then
> suspect a fault somewhere else. I just pulled out the manual, and I
> would start by lifting the connection on E7, and connecting a 10MHz
> xtal in series with a 10pF cap between the E7 pin and ground. This
> will bypass both the xtal and the frequency adjustment circuit. If
> that doesn't work, then the oscillator circuit is the problem - if it
> does work, then remove the 10 pF cap and connect the xtal directly
> between E7 and E8 - if it still works, it's almost certainly a bad
> xtal. If it worked with the 10 pF cap but not with the E8 connection,
> then check the control voltage (E9: 0.5V-17V) - if that looks good,
> then suspect the frequency control circuit. If it's bad, then try
> hooking it to +5V to verify the oscillator and then troubleshoot the
> servo board.
>
> Regards,
>
> Pete
>
>
> On Wed, Dec 21, 2011 at 5:00 AM, Azelio Boriani
> <azelio.boriani at screen.it> wrote:
>> 10MHz crystals used in OCXO are designed for the temperature of the oven
so it is not possible to find an off-the-shelf suitable crystal. Better find
a crystal coming from an OCXO.
>>
>> On Tue, Dec 20, 2011 at 7:10 PM, Peter Bell<bell.peter at gmail.com> wrote:
>>> Hi, Don
>>>
>>> It's been quite a while since I worked on a FRK, but from what I
>>> remember the only signals going to that oscillator board are power and
>>> the control voltage. It might be worth checking the control voltage to
>>> see if it's at a plausible value (nominal is 5V, spec range is 1V-16V,
>>> IIRC) - if you have a spare xtal around, just try connecting it to the
>>> pins on the PCB and see if it starts up.
>>>
>>> I'm not aware of the xtals being failure prone - I have seen one FRK
>>> with a bad one, but having seen the condition of the system it was
>>> removed from, that was clearly impact damage.
>>>
>>> Regards,
>>>
>>> Pete
>>>
>>>
>>> On Wed, Dec 21, 2011 at 1:29 AM, Don Lewis<dlewis6767 at austin.rr.com>
>>> wrote:
>>>> A little help, pls, from the more experienced. This is my first time
into one of these .I have listened and learned from you' all for some time,
though. It can be intimidating in this close-packed FRK.
>>>>
>>>> I am trying to bring up an Efratom FRK I have had for a while. The 10
MHz oscillator does not start upon power on.
>>>>
>>>> Any ideas where to look?
>>>>
>>>> I have done the following:
>>>>
>>>> * I have it opened-up and applied 24vDC and the current starts up
initially at 1.45 amp and slowly goes down to ~.6- .8 amps.
>>>> * Since I have no 10 MHz output, .so I looked at the oscillator
board and nothing out from it.
>>>> * The 17 volt dc supply-rail is good, and the VCO input signal to
the varacter control-circuitry is constantly ramping up and down slowly
between ~1volt and ~15 volts.
>>>> * All the transistors in the oscillator circuit seem ok with a
VOM,
>>>> ..and when I purposely inject a 10Mhz signal into the collector of the
'oscillator' transistor (Q3), .I do get a clean 10Mhz out of the unit (of
course it won't lock), .and I haven't even been into the rest of the unit.
>>>> * I think the crystal is bad, .I took the cover off the crystal
oven, it 'looks' ok, . and the crystal heater in the oven is heating and
regulating.
>>>>
>>>> Any ideas? .does the 10 MHz crystal in these have a history of
failing? Can I get a replacement crystal for it easily?
>>>>
>>>> How can I test this crystal for pass/fail?
>>>>
>>>> Thanks, .just my first test of the waters as a fledging time-nut.
>>>>
>>>> Don Lewis
>>>>
>>>> Austin, TX
_______________________________________________
time-nuts mailing list -- time-nuts at febo.com
To unsubscribe, go to
https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
and follow the instructions there.
More information about the Time-nuts_lists.febo.com
mailing list