[time-nuts] HP 10811-60111 - successful repair of NTC damage

jmfranke jmfranke at cox.net
Sat Jan 7 15:54:39 UTC 2012


Congratulations on a job well done and a saved prize piece of technology.

John  WA4WDL

--------------------------------------------------
From: "Frank Stellmach" <frank.stellmach at freenet.de>
Sent: Saturday, January 07, 2012 9:17 AM
To: <time-nuts at febo.com>
Subject: [time-nuts] HP 10811-60111 - successful repair of NTC damage

> Dear time-nuts,
>
> the 10811, sitting in my HP5370B, showed an intermittent heater
> circuitry failure.
> It turned out, that the NTC temperature sensor was damaged (1982 vintage).
>
> At room temperature it was conducting (100kOhm), but became open during
> heat up phase, eventually conducting properly at around 82°C, but
> failing from time to time, i.e. the star symbol on the 5370B turned on
> frequently, and the reference frequency fluctuated strongly.
>
> Normally this is a total damage of this OCXO, as the NTC is epoxied
> completely into the oven mass, and this whole assembly would have been
> to be replaced, according to the manual.
> I doubt, that it's available any more, but definitely not for a
> reasonable price.
>
> I found a compatible NTC replacement, regarding diameter (2,41mm max)
> and electrical parameters, i.e. Rn(25°C) = 100kOhm, B = 4540K, and an
> even better accuracy of 1%, i.e. 8.6kOhm at  82.8°C +/- < 1°C.
> It's from epcos, S57867S104F140. Visit epcos.com for data sheet and R(T)
> table.
>
> This NTC has 50mm short and blank wires; I couldn't get the PTFE
> insulated S861.. type.
> So the wires had to be insulated with a heat shrink tube, and the NTCs
> wires had to be assembled straight, without bending them 180 degrees
> through the 2nd hole in the oven mass. That's also much better regarding
> reliability.
>
> After complete disassembly and carefully milling out the old NTC and the
> epoxy with a 2.5mm drill,  I applied thermal grease at the new NTCs
> head, and fixed the wires with a small drop of epoxy at the other side
> of the oven mass.
>
> The reassembly was a little bit delicate, due to the short wires, and
> because the insulation of the wires was thicker than before. But
> finally, the heater circuitry stabilized normally within 10 minutes.
> Obviously, the thermal gain was not perceptibly affected by this
> slightly different assembly.
>
> The OCXO now works fine since two weeks, and its drift is in the order
> of < 1e-9 again.
>
> Frank
>
>
>
>
>



> _______________________________________________
> time-nuts mailing list -- time-nuts at febo.com
> To unsubscribe, go to 
> https://www.febo.com/cgi-bin/mailman/listinfo/time-nuts
> and follow the instructions there. 






More information about the Time-nuts_lists.febo.com mailing list