[time-nuts] FE-5680A worked for only 5 minutes.
Paul F. Sehorne
paul at sehorne.org
Fri Jan 20 19:09:45 UTC 2012
I powered up the second unit. It has been running just fine for an hour
now. The only difference I noted is that the initial current was only
1.4 amp vs 1.8 on the other one. However, I set my power supply to 16.0
volts; no other difference. After warmup the current settled at ~700 ma
and the 5vdc current at ~80 ma.
On 1/19/2012 1:15 PM, Paul F. Sehorne wrote:
> I received my two FE-5680As late yesterday. This morning I hooked one
> of them up. It locked on in a couple of minutes and at about five
> minutes quick working. There is a smell of something burnt inside. I
> have not opened the case to investigate the damage.
>
> I had the unit attached to two current limited bench power supplies.
> One power supply was set at +16.1 vdc and the other at +5.0 vdc. When
> the trace on the scope went away I noticed the +15 vdc power supply
> had gone into current limit mode at 2 amps. The +5 vdc current had
> dropped from ~80ma to about ~60. Before the failure the +15 vdc supply
> showed ~1.8 amp during the first minute or two and then dropped to
> ~700 ma.
>
> While it was working it worked fine and was very near my GPSDO in
> frequency. The scope trace was moving 5ns in about 20 seconds. I got
> no further than this when the unit stopped working. I have not
> powered up the second unit yet; preferring instead to contact
> nichegeek first.
>
> Paul