[time-nuts] A counter for phase measures
Bob Camp
lists at rtty.us
Sat Nov 3 23:06:48 UTC 2012
Hi
The relay *can* be replaced. Not so much with the input amp. If it goes, you lost the counter. The displays also die from time to time. Like the relay, they can be replaced.
Bob
On Nov 3, 2012, at 6:14 PM, Magnus Danielson <magnus at rubidium.dyndns.org> wrote:
> On 11/03/2012 09:36 PM, Bob Camp wrote:
>> Hi
>>
>> In the case of the 53132, the power supply seems to be the weak link. Out of maybe a hundred or so in the fleet, we see maybe one or two die each month. On the SR620 the power supply also seems to go from time to time. Both have the normal keyboard and display issues, but those can be fixed. If you go back to things like the 5335, the weak point is the input amp, it blows if you get +5 on it. Like power transformers - not a replaceable item...
>
> Another weak point on the 5335 is the relay. We had to replace it and the relay-holder, but once that was done, it was back up operational.
> The 5335 ticks in as the most human-friendly of the counters at work, while the 53132 is competing with the 5372 as being the most human-unfriendly, where the 5372 has more capabilities to present, so it gets used more.
>